Not sure about the newer models, but for the Delta Plus XL the H3
determination is not linear under 5V according to the manual
Harlow, Benjamin wrote:
>
> Effie,
>
> In the method you used, under “Peak detection @H2”, you will see the
> starting value of the H3 factor. Here you can select overwrite and put
> in a starting value of 5. You could try this using the reevaluate
> option for the CF document in question. But I think you should use a
> broader span of voltages to calculate the H3, like 1-7V for example.
> That way you are giving isodat more of slope to work with, thus
> reducing error in the estimate of the H3.
>
> Isodat will start with the last H3 calculated, so you need to force it
> to start at a more realistic starting point. I have one method I only
> use for calculating the H3 where I overwrite or define the H3 starting
> value to be 5.
>
> Hope this helps,
>
> Ben Harlow
>
> Washington State University
>
> ------------------------------------------------------------------------
>
> *From:* Stable Isotope Geochemistry [mailto:[log in to unmask]]
> *On Behalf Of *effie
> *Sent:* Monday, April 16, 2007 12:40 AM
> *To:* [log in to unmask]
> *Subject:* [ISOGEOCHEM] The "start value" of the H3 factor
>
> Hi all,
>
> I have started to look into dD analysis, and I have some problems
> concerning the H3 factor. I hope you can enlighten me with some
> explanations. The analysis is of water using the TC/EA and DeltaplusXP
> MS. Back ground levels are of 10 for mass 3, and 30 for mass 2. During
> the determination of the factor (that is preceded by the zero on off
> measurement while increasing H2 ref gas pressure. Increasing being in
> the intensity range of 1000mV to 2000mV), the software (Isodat 2.0)
> determines a value from which it starts to calculate the H3 factor. On
> screen it will show it as "start value". The start value is of 42, the
> H3 factor which is determined is around that figure (with a standard
> deviation 0.6).
>
> My questions:
>
> What is that "start value" and how it is determined?
>
> Assuming my factor should be less then 10, what may cause a larger figure?
>
> Thanks in advance,
>
> Effie
>
> *Effie Eliani
>
> **Stable isotopes lab
>
> ***Geological Survey of Israel
>
> This mail was sent from mail.pineapp.net
>
> ************************************************************************************
> This footnote confirms that this email message has been scanned by
> PineApp Mail-SeCure for the presence of malicious code, vandals &
> computer viruses.
> ************************************************************************************
--
Jonathan D. Karr, Ph.D.
Technical Director, Duke EnVironmental stable Isotope Laboratory (DEVIL)
Duke University Dept.of Biology & Nicholas School of the Environment and Earth Sciences
Phytotron Bldg, 71 Science Drive
Durham, NC 27708-0340
(919) 660-7418
Fax: (919) 660-7425
[log in to unmask]
laboratory website: http://www.biology.duke.edu/jackson/devil/
Adversity is an opportunity
|