One more potential story/hypothesis/solution to the HV tripping
problem. It is possible/likely there are multiple causes of what so
many of us seem to have experienced:
Several months ago our Delta Plus XP (with two source pumps, generally
running at 1.6e-6 mBar in CF mode) began tripping to 0.34 kV with
increasingly alarming frequency (increasing to an "event" every several
days). This led to a decision to advance by several weeks the annual
maintenance to clean the source. Then the filament burnt out,
requiring immediate maintenance. So I (1) replaced the filament, (2)
cleaned the source thoroughly (including scrubbing all metal parts,
filing and annealing ceramics), and (3) replaced a burnt out source
heater bulb (see Marion's comment), and (4) pump maintenance, etc. The
problem has not recurred for 6 months now. I suspect it will again
when the source is sufficiently dirty/wet? to arc across ceramic
insulators. I think in this case it was simply a dirty/wet source,
which should not be ruled out entirely as the cause in some cases.
Jonathan Wynn.
On 2 Dec 2005, at 08:27, Marion Hoch wrote:
> Hey Linnea,
>
> we have the same problem every three or four months. The HV turns off.
> You can restart it but after a few hours it trips off to 0.34 KV
> again. The first time it happened we checked our power supply system
> for nearly one week. But we could not find any fault although the HV
> turned off several times during that week. But now we have a very
> simple explanation. We found that one of the heaters did not work any
> more. We changed the heaters and the system was running without any
> problems. So the last two or three times when the HV turned off I
> checked the heaters and in all cases one of them were broken. After
> changing the heaters we had no problem anymore with the HV. This might
> be a very simple solution for your problem but in our case it works.
>
> Good luck
> Marion
>
>
Department of Geography and Geosciences
Irvine Building, University of St. Andrews
St. Andrews, Fife, Scotland, KY16 9AL
United Kingdom
phone: +44 (0)1334 463936
fax: +44 (0)1334 463949
e-mail: [log in to unmask]
|
|
|