Date:
Tue, 7 Mar 2006 09:11:09 +0100
Content-type:
text/plain; charset=ISO-8859-1
MIME-Version:
1.0
Content-Transfer-Encoding:
8bit
|
Hi Penny
we experienced a similiar increase in mass 46 background when
analysing N2O in our Precon-delta-XL system a few weeks ago. We
suspect a contamination on the Poraplot-Q column of the Precon.
Baking out the column in the GC oven at 200°C did not solve the
problem.However, after removing the column and baking in another
oven at 250°C the mass 46 background is fine.
We suspect that contamination of our sample vials with traces of oil
vapor might have caused the problem. Some of our sample vials are
pre-evacuated using a vacuum stand and we suspect that the zeolite
trap of the pump had been exhausted. The problem occured when
running large ( 500 mL ) gas samples. We are still not certain if our
interpretation is correct. Therefore we would be interested to know if oil
vapor contamination would also be a possible explanation to your
mass 46 problem.
Cheers,
Reinhard
Date sent: Mon, 6 Mar 2006 09:57:19 -0500
Send reply to: Stable Isotope Geochemistry
<[log in to unmask]>
From: Penny Higgins <[log in to unmask]>
Subject: [ISOGEOCHEM] High Mass 46 background
To: [log in to unmask]
> Hi all,
>
> We've been running carbonates (CO2) on our Delta Plus XL, via the
> Gasbench and PAL, for nearly a year now. Everything has been running
> smoothly. Alas, in the past few weeks I've noticed that the mass 46
> background has increased dramatically from 10-20mV to nearly 50mV
> (sometimes more). Our standard deviation on d18O (based on ten peaks)
> has increased to 0.15 or much higher. Carbon is still running very
> well, and I am certain that water in the sample is not getting to the
> mass spectrometer source. We check linearity every day before running
> analyses, so I know that the instrument is performing magnificently
> and is linear. The mass 46 background on the reference peaks is
> generally much lower than that of the sample peaks, though these days
> still tops 20mV.
>
> Does anyone know where the excess mass 46 is coming from. There
> appears to still be a clear separation between the nitrogen peak and
> the carbon dioxide peak coming out of the GC column. However, I wonder
> if the GC column may yet be bad. We did once draw phosphoric acid into
> the GasBench (a procedural problem we have since fixed). I wonder if
> this may have damaged the GC column?
>
> Any other thoughts?
>
> Thanks,
>
> ~Penny
>
> *******************************************************************
> Dr. Pennilyn Higgins
> Research Associate
>
> "SIREAL"
> Stable Isotope Ratios in the Environment Analytical Laboratory
>
> Department of Earth and Environmental Sciences
> University of Rochester
> 227 Hutchison Hall
> Rochester, NY 14627
>
> [log in to unmask]
>
> Office: 209b Hutchison Hall Lab: 209 Hutchison Hall
> Voice : (585) 275-0601 Outer lab: (585) 273-1405 FAX :
> (585) 244-5689 Inner lab: (585) 273-1397
>
> http://www.earth.rochester.edu/SIREAL/index.html
> *******************************************************************
>
______________________________________
PD Dr. Reinhard Well
Institut für Bodenkunde und Waldernährung
Universität Göttingen
Büsgenweg 2
37077 Göttingen
Germany
phone +49 551 395507
fax +49 551 394619
|
|
|