Dear Erik,
To come up with a satisfactory and useful answer for you, can you tell me a
bit more about this (off list if you prefer)?
What instrument and, hence what type of combustion reactor are you using?
When you re-oxidised the reactor, at what temperature and for how long?
Best,
Wolfram
On Apr 23 2007, Erik D. Pollock wrote:
> List Members,
> I am having an issue with the chromatography going bad after
> oxidizing the combustion column. Prior to oxiding the column our multi
> fame house standard has peaks that range from 700 mV to 1300 mV with peak
> widths of about 20 seconds. After oxidation the same sample has peaks
> which are less than 100mV to 400mV and widths 35 to 50seconds. The
> problem is unaffected by GC service, column cutting, new inlet liner, new
> septa. The unsatisfactory solution is make many standrad injections which
> slowly improve the chromatography. Has any one had similar issues. I have
> heard rumors of some folks sending part of a solvent peak through the
> combustion column? Our GC-C-IRMS set up for FAMEs It is an HP6890 with a
> Delta Plus and a GCIII combustion interface. The MS seems operating
> normally, good backgrounds, good refernce peaks, little to no change in
> these before or after oxidation.
>
> Erik D Pollock
> University of Arkansas Stable Isotope Lab
> 632 Science and Engineering Hall
> Fayetteville, AR 72702
> 479-575-4506
> http://www.uark.edu/ua/isotope/index.php
>
|
|
|