Maybe BAKE OUT the GC Oven of the EA?
Raise the oven Temperature to 110 deg C for 4 hours, then return it to
your regular operating T (55C, or 60C, or whatever it is). Our EA
manufacturer recommends doing this every 3-4 months.
--
Charlotte Lehmann
Assistant in Instruction
Bates College
Department of Geology
206A Carnegie Science Building
44 Campus Avenue
Lewiston, ME 04240
Phone: 207-786-6485
FAX: 207-786-8334
Quoting "Stephen S. Howe" <[log in to unmask]>:
> Isogeochemists:
>
> I am having a new problem with unacceptable peak tailing during carbon
> isotopic analyses of samples using our Carlo Erba NA 1500 Series 2 NC
> elemental analyzer interfaced with our Optima mass spectrometer.
>
> Last week, I finished the analysis of about 700 samples, about 600 of
> which were marine sediment in 5x9 mm tin capsules, followed by about
> 100 samples of marine mammal muscle tissue in 3.5x5 mm tin capsules.
> Chromatography was excellent during these runs, with just the faintest
> hint of tailing as the CO2 peaks came down to background only when the
> quartz inserts needed changing as they filled with ash and the sample
> drop position started moving slightly above the optimum hot zone,
> stretching the duration of the combustion very slightly. The magnesium
> perchlorate water trap was changed after about 650 samples, and the
> remaining 50 or so analyses were fine. I decided to change the
> oxidation and reduction chemicals in their respective furnace tubes
> after what appeared to be NO started to break through during the runs a
> second or two after the time N2 would be expected (even though the mass
> spec was tuned only to CO2 during these runs). Sure enough, the copper
> turnings in the reduction tube were completely oxidized.
>
> I brought the reaction tubes back down to room temperature to change
> the chemicals, and replaced both the oxidizing and reducing chemicals
> with new chemicals of the same vintage as the chemicals used for the
> 700-sample run. I loaded everything the same way I always do, taking
> pains to insure dead volume was at a minimum. Leak tests were made and
> flow rates were set at room temperature and then both again at high
> temperature. He, O2, and CO2 supplies are from the same cylinders as
> before, at the same delivery pressures as before both at the EA and the
> reference gas injector box. Capillary positions remain unchanged and
> mass spec background when tuned to N2 and ion gauge readings are as
> good as ever. Stability runs are excellent.
>
> Samples of the peach leaves standard (I usually run at least a dozen
> or so after changing chemicals) appear to be combusting
> beautifully, with a bright flash two seconds after the drop, as I
> usually set. However, the chromatography is not acceptable,
> exhibiting pronounced tailing as the CO2 peak attempts to return to
> baseline, both for the TCD and the mass spec traces. Sometimes
> this tailing extends for at least 60 seconds beyond the time it
> used to reach baseline. The shapes of the 2/1 and 3/1 ratio traces
> are also noticeably affected. Extending the run 60-120 seconds
> past the usual length of 360 seconds allows the baseline to drop
> some more between samples, but I'm not able to get any consistency
> among isotopic values, even with adjusting the HS open timing to
> optimize the acquisition window. I'm getting values of -23 ± 2 per
> mil, instead of my customary -25.9 ± 0.2 per mil.
>
> It seems logical that dead volume is to blame for causing the tailing
> on both the TCD and mass spec traces, since combustion is good,
> especially as the only thing I've changed are the chemicals in the two
> furnace tubes, but I really can't believe my packing was poor enough to
> cause this amount of tailing. I really don't want to bring the EA back
> down to room temperature and repack the tubes again (let alone waste
> the new and costly chemicals I just put in). Do any of you experts
> have any suggestions for something I might have missed or might try?
>
> Thanks for any help anyone can give.
>
> Steve
> --
> ---------------------------------------------------------------------------
> Stephen S. Howe Office: (518) 442-5053
> Dept. of Earth & Atmos. Sciences Stable Isotope Lab: (518) 442-4471
> Earth Science 352B FAX: (518) 442-5825
> University at Albany E-mail: [log in to unmask]
> 1400 Washington Avenue http://www.albany.edu/geosciences/sshowe.html
> Albany, NY 12222-0001 http://www.albany.edu/geosciences/sirmslab.html
> ---------------------------------------------------------------------------
|