Have you also tried the cleaning out the dropper and the tubing with that fine stiff wire? The acid reservoir vent stopcock isn't plugged is it? Seems like you should be able to clean rather than purchase. Good luck.
andy
Andrew Schauer
ISOLAB
Department of Earth and Space Sciences
University of Washington
Seattle, WA 98195
206.543.6327
[log in to unmask]
----- Original Message ----
From: tcp121 <[log in to unmask]>
To: [log in to unmask]
Sent: Thursday, May 8, 2008 2:31:03 PM
Subject: Re: [ISOGEOCHEM] acid valve capillary for KIEL IV
Hi Million...
Have you tried advancing the acid tubing at all? We find that over
time the acid tubing at the valve eventually loses its elasticity and
does not expand when the valve itself is opened. When this happens we
remove the tubing from the valve and cut away the collapsed portion,
then advance the tubing so that a fresh part is within the valve.
This way you can greatly extend the life of the acid tubing, until it
becomes too short to advance it further.
Tim
On 8-May-08, at 3:24 PM, Million Hailemichael wrote:
Hi all,
Does anyone knows an alternative supplier other than Therm for an acid
valve
capillary (part #0598651) for KIEL IV Carbonate device. I checked
Thermo
and they are very expensive almost $360.00 only for the capillary. The
sequence terminates after few runs because of timing out when acid
droping.
I tried increasing the acid drop time from 120 sec to 180 sec, it was
fine for
few runs and start terminating the sequence again.
Thanks
Million
-------------------------------------------------