Hi Chris,
We are able to run the GC Pal with either the Gas Bench or TCEA water
without changing the definition of TRAY01. We also don't use a Job. The
Job Queue is empty. I'm not sure exactly why it works. I'm still a
little baffled as to how Isodat tells the GC Pal which tray to go to.
In any event, here are the specifics:
Isodat Configurator:
Conflo Water
MS Delta Plus XP
Capillary
Conflo II Interface
Input
Elemental Analyzer
Sampler
ConFlo- Liquid
GasBench
MS Delta Plus XP
Capillary
Gas Bench
Input
A200S
Isodat Methods (both Conflo Water and GasBench)
Time Events
Acqu Start - Immediately
Isodat Sequence (Conflo Water)
AS Sample column - Vial #
AS Method column - >Internal No5
Isodat Sequence (GasBench)
AS Sample column - Vial #
AS Method column - >Internal No8
GC Pal
Setup
Objects
Trays
TRAY01 - VT98
TRAYGB - GB-8*11
GC Pal
Methods
A200S-5
Inject in - GC Inj1
GC Pal
Methods
A200S-8
Inject in - none
(Adjust timing using Fill Strokes and Pullup Del)
Good luck,
Georges
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Georges L. Paradis
Analytical Lab Manager
Marine Science Institute
Bldg 520, Room 4013
University of California Santa Barbara
Santa Barbara, CA 93106
805-893-7182 (ICPMS lab)
805-893-3802 (MSI lab)
[log in to unmask]
http://analab.msi.ucsb.edu/
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Linnea Heraty wrote:
> Hi Chris
>
> There is a "secret" to accessing the advanced controls of the PAL.
> Before selecting Setup from the main menu, press F3. Then when you
> select Setup you will see several more sub-menu options. Select
> Objects, then Trays and the copy tray function should be an option.
>
> I have a library of different trays saved on the hand held controller
> and it is a quick switch-o-change-o to copy the appropriate tray to
> TRAY01. Just be careful who knows about the power of F3, much chaos
> can ensue.
>
> However, I have yet to get TRAYMAKECOFFEE to work, the coordinates of
> the tray holder must be off....
>
> Linnea
>
>
> At 06:35 AM 9/4/2008, you wrote:
>> Thanks everybody for your responses.
>>
>> I do not appear to see any function within the handheld PAL
>> controller to enable a copy/name trays or tray holders, just to
>> delete them. However, I was able to open a pre-installed tray holder
>> object using PAL's object manager (in a text app) and change any
>> reference of TrayX to Tray01. I can then upload this object to the GC
>> PAL and it works.
>>
>> Obviously, this is not completely ideal as we now have to upload
>> through "PAL loader" backups each time we want to change peripherals.
>> But it works, so I won't complain.
>>
>> Cheers again,
>>
>> Chris
>>
>> --
>> Christopher Wurster, Ph.D.
>> School of Geography and Geosciences
>> University of St Andrews, Irvine Building
>> St Andrews, Fife KY16 9AL
>> Scotland, UK
>>
>> Office: 510
>> Office phone: 01334 463936
>> Lab phone: 01334 463916
>> Fax: 01334 463949
>> website: http://www.st-andrews.ac.uk/~cmw10/Wurster/Home.html
>> <http://www.st-andrews.ac.uk/%7Ecmw10/Wurster/Home.html>
>>
>>
>>
>>
>>
> Linnea Heraty
> Department of Earth and Environmental Sciences
> University of Illinois at Chicago (MC-186)
> 845 West Taylor Street SES 2438
> Chicago, Illinois 60607-7059
> office/lab: 312-413-0098
> fax: 312-413-2279
> email: [log in to unmask]
>
|