Hi John.
I guess it is not the material of this ferrule, you can use PEEK or GVF
like you want.
But the modification is the core through the ferrule.
First, I replaced the valve seat with an original GVF from SGE.
It was a bad performance, half signal height and so on.
So I and my colleague asked a different technician and we get nearly the
same answer.
Thermo use the GVF from SGE with a bigger core of 0.8 mm.
So I made the core of our GVF bigger and get a good performance now.
Best,
Sven
John Hill schrieb:
> Sven
>
> This is interesting. I replaced the SGE needle valve a few months back
> on our Delta V. SGE have modified the needle valve and it uses PEEK
> ferrules whose ID must be around 0.4 to 0.5mm. So basically we need to
> changed back to GVF ferrules with the larger ID, correct?
>
> John Hill
> UCL
>
> Sven Steinbrenner wrote:
>> Dear Istvan,
>> there is a modification at the needle valve seat ferrule.
>> Thermo use a ferrule with a bigger core inside (0,8 mm).
>> You can order it with the pn 1221460 GVF-ferrule for Needle valve id
>> 0.8mm.
>>
>> I replaced at our 3 Deltas all seat ferrules and was wondering about
>> the bad performance.
>> Lower signal height; bad Ref gas peaks (round at the beginning).
>> After a few mails to the support, we get the info about this
>> modification.
>>
>> I hope this helps.
>>
>> Best wishes
>> Sven
>>
>> Fórizs István schrieb:
>>> Dear Ulrich and others who helped us with advises,
>>>
>>> after all we have made the first measurements during the night. Now
>>> it worked acceptable, although the peak height has not encreased (it
>>> is about half than was before). But now the H3 factor is far better.
>>> After the cleaning of the ion source the H3 factor varied between 16
>>> and 24!!
>>> Yesterday it was between 6.3 and 6.4, and looked stable. Today it
>>> was 6.15 and 6.05 (we have measured it two times), and looks stable.
>>>
>>> What we have found are:
>>>
>>> 1) One plate in the ion source was 90 degrees off. It is the closest
>>> one to the ionization box, no. 02, and without wire connection
>>> (there is a metal ring between this plate and the ionization box).
>>> When we put it to the proper position we expected a significant
>>> encrease in the peak height, but it did not happen. We do not
>>> understand why.
>>>
>>> 2) The connections of the high voltage cable (red) on the outer part
>>> of the ion source were (and are) loose. Practically after may
>>> disassembling and reassembling one of these connections was
>>> disconnected. When we discovered it, we realized that all these
>>> connections are loose a priori. So we have to be very careful when
>>> cleaning the source.
>>>
>>> What we have not checked is the needle valve on the input port of
>>> the MS. We have noticed that now it is more difficult (needs more
>>> force) to close and open it then before. May be that it causes the
>>> half height peak?
>>> What is your experience with the needle valve?
>>>
>>> Thank you very much for the help provided by many of you!
>>>