Hi Mike,
try to heat out the GC column of your EA, this helps on our EA with the
same mass 30 peak problem.
good luck,
Sven
Kubo, Michael D. (ARC-SSX)[SETI INSTITUTE] schrieb:
>
> Hi everyone,
>
> After a long hiatus (several months of down time), I am gearing up to
> run some EA samples for nitrogen and carbon on our Delta Plus and
> Carlo Erba 1108 EA and I’m noticing something quite unusual. While
> idle and with the magnet tuned for nitrogen, the background for mass
> 30 is around 80 mV, while those for 28 and 29 are approximately 20 mV
> each. During the run, when introducing the nitrogen reference gas, the
> trace for masses 28 and 29 shoot up from their background (of course),
> but what’s unusual is that the mass 30 shoots up as well, but with a
> significant (~1 second or so) lag to approximately half of the height
> of the other gases, or about 400-500 mV. When the reference gas “turns
> off”, the reverse happens, with mass 30 lagging by about a second to
> SLOWLY return to about 100-120 mV background (it never returns to 80
> during the run). After my reference gas injections, the mass 30 trace
> slowly tries to drift back towards baseline, until my nitrogen sample
> peak elutes, at which point the mass 30 trace rises again to mimic the
> trace for 28 and 29.
>
> I’m a little confused here. At first I thought it might be a spent
> reduction (packed copper) reactor, but then I thought I would only see
> this on my sample peak and not on my reference gas peaks.
> Alternatively, if there was something magically wrong with my
> reference gas, I would only see this phenomenon on my reference gas
> injections and not with my sample peak. Mass 44, 45, and 46 traces for
> carbon look totally fine.
>
> I will say that I recently had to replace the capillary in my Conflo
> II interface – the one that runs from the CO2 regulator to the
> capillary “sniffer”. Could this have anything to do with my problem?
>
> Any and all opinions would be greatly appreciated.
>
> Cheers,
>
> Mike
>
> ****************************************************************************************************************************************
>
> Michael D. Kubo
>
> NASA Ames Research Center
>
> M/S 239-4
>
> Bldg. N239, Room 327
>
> Moffett Field, CA 94035
>
> (650) 604-6110
>
--
Sven Steinbrenner
Max Planck Institute for Evolutionary Anthropology
Department of Human Evolution
Deutscher Platz 6
D-04103 Leipzig
Germany
Phone: 49 (0)341 - 35 50 388
Fax: 49 (0)341 - 35 50 399
mail: [log in to unmask]
|