Paul
We had issues with the PAL injecting samples into a GC. The issue appeared
to be that the PAL was waiting for a ready signal from the GC. Although
your PAL is not connected to a GC or waiting for a signal it may be that
the relay is set for a "ready to inject" signal. Although I don't have
access to the PAL at the moment, we resolved the problem by changing the
default values for the ready to inject signal within the global methods of
the PAL, such that it did not need a ready signal.
Mark
>Content-Type: text/plain; x-avg-checked=avg-ok-5A2721; charset=us-ascii;
>format=flowed
>Content-Transfer-Encoding: 8bit
>
>Listmembers,
>
>I'm having trouble with my CombiPal, which in the past has worked just fine
>on my TC/EA (coupled to a Delta Plus XL). Isodat seems to be unable to
>trigger an injection from a sequence run.
>
>- The Pal seems fine, I can manually trigger an injection (using method1).
>- The Loader software is able to communicate with the Pal, on Com1.
>- Isodat is also configured for the Pal on Com1.
>- The A200S method builder within Isodat is successfully able to send a new
>method to the Pal.
>- Within the sequence, I'm selecting "internal method 1" and a valid sample
>number.
>
>Does anyone have a thought as to why the software isn't triggering the
>autosampler?
>
>(I don't have the very latest Isodat patch: when I tried installing that,
>Isodat crashed every time I tried to start an autosampler run)
>