Greetings Isotopists!
I'm having a bear of a time getting an A200S liquid autosampler to
work w/ Isodat 2.0. Here's the story....
Equipment:
Delta+XL w/ TC/EA
A200S autosampler
Start Signal Source = REMOTE
Ready Active State = CLOSED (Though I've tried all four options,
CLOSED, OPEN, OCI-CLOSED, OCI-OPEN)
Method = 9
set to INJ-Ready
Isodat 2.0 - SP 2.38
Set-up: A200S Remote connected to Isodat PC COM2 w/ DB25-DB9 cable p/
n 0492310
I've tried several configurations:
Config 1: ConFlo + EA + A200S (standard choice in the ConFlo sets
section of configurator - contains the ConFlo-Liquid device in the
drill down configuration tree, not the A200S device)
Config 2: ConFlo + EA..... then manually added A200S device (from
devices) to > Capillary > ConFlo > Input > EA > Sampler. A200S
device is set to COM4
Config 3: ConFlo + EA..... then manually added ConFo-Liquid device
(from devices) to > Capillary > ConFlo > Input > EA > Sampler.
ConFlo-Liquid Device is set to COM2
Method set to "Start immediately" after intitialization
Sequence - set to AS Method #9 (default)
COMPORT settings are:
Baud=9600
Data Bits=8
Stop Bits=1
Parity=None
Append CR=CHECKED
Append LF=UNCHECKED
Scenario #1 - default ConFlo+EA+A200S config. Sequence hangs at
"Waiting for AS..." - message repeats, it's obviously waiting for an
"AS ready" signal. AS never starts
Scenario #2 - Config #2 (which contains A200S device configured to
work on COM4 instead of ConFlo-Liquid (COM2)device): Sequence doesn't
even contain the "AS Start" column. Sequence triggers EA (via GPIB/
pneumatic trigger) ONLY when COM2 is connected to working (confirmed)
PAL80 AS. If COM4 is not attached to PAL80 it doesn't get past the
"Waiting for AS..." If COM4 is attached to A200S I also get the
"Waiting for AS..." message. This seems to indicate that Isodat is
able to receive a ready signal though this is only proven on COM4,
not necessarily on COM2
Scenario #3 - Config #3 (ConFlo-Liquid device on COM2): Again,
sequence simply hangs at "Waiting for AS..." Isodat doesn't seem to
be receiving the ready signal from the A200S.
Scenario #4 - Config #2 modified so that ConFlo-Liquid Device is set
to COM4 (which is actually connected to the functioning PAL80).
Sequence proceeds fine and starts the PAL80 as expected (b/c I've
connected it to the PAL80 instead of the A200S)
I think that these results indicate two possibilities:
(1) The cable that I'm using between the A200S and the PC is
incorrect. It has different pin-outs compared to the cable between
the PC and the Pal80, that is for sure. I do not know whether or not
the pin assignments should be the same for both A200S-PC and Pal80-PC
cables. The Pal80-PC cable has a null modem config (though I haven't
figured out what type of null-modem set it has). The 0493210 cable
from A200S-PC is not a null modem cable as best I can tell, no pins
are jumpered. It's very possible tha the cables SHOULD be different
- I just don't know at this point and it is a possible cause of my
comm failure. Does anyone know the exact specs of the DB9-DB25 cable
that should be used to connect the A200S to the Isodat PC?
(2) The other possibility as I see it is that I somehow mucked up the
COM-PORT settings or the configurations. If anyone as any
recommendations regarding COM-PORT settings or configurations I'd
love to hear it.
My apologies for the ridiculously long post. So, if you're reading
this AND following it then you probably have some knowledge that
would help
Many thanks,
Ethan