Re: carbonates on Thermo Delta XP
From:
Penny Higgins <[log in to unmask]>
Reply To:
Stable Isotope Geochemistry <[log in to unmask]>
Date:
Fri, 9 Dec 2005 08:42:18 -0500
Content-Type:
text/plain
Parts/Attachments:
text/plain (111 lines)
Hi Attila,

I've noticed similar troubles for us, too. We run carbonates nearly every 
single day and drift can be VERY problematic. We run about 44 samples in a 
run (before rechecking linearity), and actually measure 10 sample peaks. 
Four things we've done to reduce the problem:

1) To save time and mental energy, we use LIMS for light stable isotopes 
software developed by Ty Coplen at the USGS to make it easy to correct for 
drift in our post-analysis corrections. This saves a lot of grief and works 
great, provided that the drift is linear, which in our case it usually is.

2) We set our Extraction 1 and Extraction 2 (aka Extraction symmetry) at 
about 40%, which helps our linearity considerably, though we lose a bit of 
sensitivity.

3) And perhaps the most important thing we do, we pre-treat all our 
carbonate samples to eliminate potential contamination from organics that 
may be in the samples. This has incredibly improved our repeatability and 
yield from samples. Plus, by getting the organics out, our nitrogen peaks 
are smaller, if not non-existent, and I'm not concerned about accidentally 
getting sulfur or other 'sticky' molecules in the instrument. This allows 
us to comfortably measure as much as 2mg of carbonate in a sample. We only 
need about 0.1mg of pure carbonate to get a good analysis. I'd be happy to 
send you our chemical pre-treatment protocol if you want it.

4) The last thing we did, which required some experimentation, was to quit 
using the Labco Exetainers. We now use a 4mL tube and caps with a silicon 
septum lined with teflon toward the inside of the tube. The troubles with 
the exetainers were, a), the opening at the top of the cap was too narrow 
to insure that the sample needle did not pierce too close to where the acid 
needle pierced, and b), the septum material of the exetainers actually 
reacted with acid to generate a gas. We had continual troubles with drops 
of acid remaining on the inside surface of the septum, then acid getting 
drawn into the GasBench and making a mess of everything. To further 
alleviate this problem, we actually stopped adding acid and sampling in the 
same sequence. We now add acid first (either using the acid pump or by 
hand), mix the acid and sample using a vortexer, then return the samples to 
the GasBench for analysis. I can send you a more detailed explanation of 
our protocol if you want.

As is the case for you, waters on the GasBench run great for us. Carbonates 
are just difficult.

I hope this is of use to you. ...and if anyone has any comments on what 
we've done, I'd also be happy to hear them.

Cheers,

~Penny

At 05:31 AM 12/9/2005, you wrote:
>Dear All,
>
>We struggle with a strange phenomenon with our new delta + XP & Gasbench 
>III. It worked perfectly with carbonates, but about a month ago the d13C 
>and d18O values started to shift during analyses. In a set of 40 samples 
>(including standards) the d13C and d18O values shifted by about 1 permill, 
>in opposite directions (d18O positive, d13C negative). Then, the shift 
>turned to a sinusoidal drift, again C and O isotopes moving in opposite 
>directions. We checked the system with pure CO2, the analyses were 
>perfect. We changed the acid, no improvement. After a set of analyses, the 
>samples were re-analysed the next day, the shift was much less significant 
>(about 0.3 permill, nice linear, correctable).
>
>The system was tuned to linearity, and there is no correlation with sample 
>size.
>Originally we analysed 7 sample peaks, then changed to analysing 4 peaks 
>(to reduce GC & Nafion saturation). No improvement, the analyses are 
>getting worse and worse.
>
>d18O and dD on waters (equilibration with CO2 and H2) work nicely, so the 
>problem is related to the carbonates.
>
>Can you give any hint what the cause of this strange behaviour may be?
>Thanks a lot in advance.
>Attila
>
>--
>*********************************************************************************
>Dr. Attila Demeny
>Geokemiai Kutatointezet - Institute for Geochemical Research
>Magyar Tudomanyos Akademia - Hungarian Academy for Sciences
>Budapest, Budaorsi ut 45, H-1112, Hungary
>Tel: (361-)309-2600/ ext. 1223 (office), 1394 (lab)
>Tel/fax: (361-)319-3137
>e-mail: [log in to unmask]
>Http://www.geochem.hu/
>*********************************************************************************

*******************************************************************
                       Dr. Pennilyn Higgins
                        Research Associate

                             "SIREAL"
   Stable Isotope Ratios in the Environment Analytical Laboratory

         Department of Earth and Environmental Sciences
         University of Rochester
         227 Hutchison Hall
         Rochester, NY 14627

                    [log in to unmask]

Office: 209b Hutchison Hall               Lab: 209 Hutchison Hall
Voice : (585) 275-0601              Outer lab: (585) 273-1405
FAX   : (585) 244-5689              Inner lab: (585) 273-1397

           http://www.earth.rochester.edu/SIREAL/index.html
*******************************************************************

ATOM RSS1 RSS2

LIST.UVM.EDU CataList Email List Search Powered by LISTSERV