| Subject: | |
| From: | |
| Reply To: | |
| Date: | Wed, 30 Sep 2009 12:35:23 -0700 |
| Content-Type: | text/plain |
| Parts/Attachments: |
|
|
Aaron,
You did not specify what type of gas mixture you are separating. If the
samples has O2 and Ar in it, then these will elute from the standard EA
GC column at the same time and interfere with the N2 peak. Normal
procedure is to inject before the Cu column to remove the O2, then
through the water and CO2 traps to remove those gases. Even so, it
might still be better to then use a molecular sieve column to separate
any trace O2 and the Ar from N2, rather than use a standard CN column.
Paul Brooks.
Aaron Shultis wrote:
> Isotope geochemists
>
> I have heard of labs installing a gas port between the furnace and the
> GC on an EA. I briefly checked the archives under "gas injector" and
> "gas port" but did not find what it is that people are installing.
> I have an Eurovector set up for N and C. I will be injecting a gas
> mixture with a small amount of nitrogen (which is what I will be
> measuring)
>
> Any ideas out there?
>
> Aaron
>
> --
> Aaron Shultis, Research Scientist
> Water Center
> School of Natural Resources
> University of Nebraska-Lincoln
> 103 Water Sciences Lab
> Lincoln, NE 68583-0844 U.S.A.
>
>
--
Mr. Paul D. Brooks,
Dept. Integrative Biology MC3140,
3060 Valley Life Sciences Building,
UC Berkeley, 94720-3140.
[log in to unmask]
phone (510)643-1748
FAX (510)643-1749
http://ib.berkeley.edu/groups/biogeochemistry/
|
|
|