Hi all,

We had some
discussion about the LGR DLT-100 syringe at this years Continuous Flow
conference. Since returning I believe I have made a bit of progress (at
least for our instrument) that might be useful to other LGR DLT-100
users. I called Hamilton asking about the 1.2 uL syringe (p/n
203185/01) recommended for the LGR liquid water instrument. They sent
me a care and maintenance pdf and a precision and accuracy statement
pdf. Those are posted here:

http://depts.washington.edu/isolab/IsoGeoChem/

Note the tiny differences in injection volume that the LGR picks up
that we are concerned about are well below any precision specs that
Hamilton claims. Also, we have quite a pile of "old" syringes that once
gave good injection volumes. According to Hamilton, these syringes are
serviceable. They recommend an alcohol rinse. I have made bad syringes
start working again by doing the following with a method created for
the PAL autosampler:

1) rinse 10 times with reagent grade acetone using a reasonably SLOW plunger motion
2) rinse 10 times with reagent grade methanol or ethanol (they both
have worked...even the alcohol mix has worked), again, with a very slow
plunger motion
3) rinse 10 times with DI water (ok we use 18 M ohm, but DI should be fine)
4) in a separate PAL method, we then do 300 actuations of the plunger
with the syringe in DI water, here again, with SLOW plunger motion

After doing 1 through 4 , it takes 2-3 injections for the H2O_cm_3 to
return to 3.2*10^16 but then it usually stays there. In addition to
this process, we do #4 (above) if the instrument has sat idle for more
than a few hours and we start and end each run with a vial of 18 M ohm
water. Note, all of this assumes the rest of the instrument is
functioning properly (pump, new septum, fresh reagents, etc etc).