Hi Tot
It looks to me that you have 14 volt (14000 millivolt)
collectors on the deltaplus. If this is so than you are maxing out
the collectors and when this happens the computer just quits taking
info from the collector(s) until the signal is back below saturation.
Reduce sample size and the problem will be solved. Jason