Hi, Sue,
The 940C @ the oven may be high enough for organic compounds. As Paul
suggested, O2 should be enough in your combustion oven, otherwise the
later eluted compounds may not have enough O2 to be completely combusted
and that will result the depletion in 13C for your later RT compounds.
You can try manual integration, but if the peak shape is good, the
result may be as the same as the auto integration. By the way, are you
using the Finnigan or Isoprime IRMS?
Wendy Zhang
Air Quality Research Division
Environment Canada
4905 Dufferin Street
Toronto, Ontario M3H 5T4
Tel: (416) 739-5792 (office)
Tel: (416) 739-5822 (lab)
Fax: (416) 739-4664
Email:[log in to unmask]
-----Original Message-----
From: Stable Isotope Geochemistry [mailto:[log in to unmask]] On
Behalf Of Susan Ziegler
Sent: April 28, 2006 11:52 AM
To: [log in to unmask]
Subject: Re: [ISOGEOCHEM] GC-C-IRMS issues
Thanks Paul:
I have cleaned the injector per manufacturers recommendation, but
omitted the brushing step since it didn't look too bad. So perhaps it
needs a more rigorous cleaning.
It does seem as though the combustion column is to blame. I did get
two decent runs of this standard before observing mass dependent
fractionation. Prior to these last tests the new combustion oven was
oxidized at 940C for 30 minutes followed by overnight at 540C. I'll
go ahead and reoxidize at each temperature longer. Perhaps the
reaction at 940 was not complete enough.
Thanks again for your input!
Sue
On Apr 28, 2006, at 11:40 AM, Paul Eby wrote:
> Sue,
>
> Two possibilities come to mind: your combustion oven, while new,
> might not have enough oxygen. As the run progresses, oxygen is
> depleted and results get worse.
>
> The other potential source might be your injector. If it isn't hot
> enough, or it's dirty, you could get a mass dependent fractionation.
>
> Paul Eby
> University of Victoria
>
>
> At 09:29 AM 28/04/2006, you wrote:
>> Hello All:
>>
>> I am seeking any possible hints to solve a problem that has developed
>> in our GC-IRMS system.
>>
>> At the moment the suite of standards we run for d13C of FAMEs in
>> order to check for consistency no longer exhibits a constant
>> correction factor across all the compounds in each standard
>> chromatogram which range from 14:0 through 23:0 in size. In fact we
>> are seeing greater depletion in 13C for the longer (or later RT)
>> FAMEs by as much as a few per mil relative to the first compounds
>> (14:0). I have been trying to isolate the variables and this is what
>> I've found so far:
>>
>> 1. Flow is constant throughout the entire GC run (2 mL/min). 2. CO2
>> injected throughout the run remains constant in it isotopic
>> composition so does not exhibit the depletion see in compounds that
>> must be combusted. 3. I figured it must be the combustion column so
>> it has been replaced and we are again seeing the same problem.
>> 4. Peak shape looks pretty good with widths remaining within 15-20s
>> and fairly sharp ratio traces. The ratio traces do become a little
>> broader with the last few more deplete compounds but not particularly
>> worse than runs where we've gotten good consistent values.
>>
>> We are using an Agilent (HP) 6890 GC with a split/splitless injector
>> in splitless mode. I have changed the septa (done frequently
>> anyway), liner, and the gold seal and washer. The GC column is very
>> new but also has been cut (`0.5m) to see if reactive sites may have
>> been a problem. We are using the metal Valco cross piece and have
>> been now for years without a problem, could this have developed
>> active sites that may yield such symptoms?
>>
>> Thanks for any information anyone might have regarding this
>> problem...I've been at this for some time now and am in danger of
>> pulling out all my hair!
>>
>> Best regards,
>> Sue
Susan Ziegler
Associate Professor
Department of Biological Sciences
University of Arkansas
Fayetteville, AR 72701
479.575.6944 (lab 6342)
479.575.4010 (fax)
http://www.uark.edu/~susanz/
|