Dear all,
Our Finnigan Delta Plus XL mass spectrometer is connected to an EA and
a TC/EA via a Finnigan MAT Conflo III split interface. We ran sulfur
isotopes last week, and everything went alright until we switched back
to CN. When all the
references are turned off, mass 28, 29, and 30 backgrounds are very
high, (~ 3-3.5 V).But mass 44 and 2 are normal (<0.01 V for mass 44,
and ~0.2 V for mass 2). The high background goes to normal when the
needle valve, where the gas comes into the Delta Plus, is turned off.
It looks like air is getting into the He somewhere upstream of the
Delta Plus needle valve. Our EA and TC/EA have no leaks, neither do the
needle valve nor the joints where gases come in and out of the Conflo
box. The gauges on the ion sources are normal (10^-3 and 10^-6 mbar
with He flowing). I double checked the collector and the resistor on
Channel 3 is set properly. We switch this resistor to run sulfur
isotopes regularly every year, and haven¡¯t had trouble before. The
Helium tanks are still the same ones we used before and during the
sulfur isotope run.
So if anyone here knows what is going on, or have any suggestions,
could you please let me know? Thanks a lot.
Majie Fan
Department of Geosciences
University of Arizona