HI Rheane--
You might be having problems with the reactor holding onto its
conditioning. Though I prefer running a TC/EA at 1450, when I was doing
GC-TC the temp was at 1400. When you run above 1400, the permeability
(be it micro-fissures or diffusion) of water/air thru the ceramic
noticeably increases--and you'll see a rise in the baseline. For a GC-TC
you can burn thru the carbon coating on the inside of the ceramic tubing
and lose some of the reducing conditions. What might be going on is
that the first peak is, in effect, re-conditioning the reactor for the
latter peaks. You can try re-conditioning the reactor. You may add a
short alkane to the sample mix--one that elutes early and that you let
thru the reactor (with sample dilution on) to condition the reactor.
Eventually the reactor will die when it's leaking faster than it can be
conditioned....
good luck,
gerry
On 2/24/2016 5:55 AM, Wolfram Meier-Augenstein (aps) wrote:
> Hi Rheane,
>
>
> >From what you describe it would seem your observations may be sample size related.
>
> To obtain reproducible d2H values on a GC-IRMS system peak heights should ideally be >3V.
>
> Another thing to look at would be your parameters for peak detection, integration and background correction to see if changes there might make a difference.
>
>
> Best,
>
> Wolfram
>
> ******************************************************
> Prof. Dr W Meier-Augenstein, CChem, FRSC
> Stable Isotope Forensics & Analytical Sciences
>
> Robert Gordon University
> School of Pharmacy and Life Sciences
> The Ian Wood Building
> Garthdee Road
> Aberdeen
> AB10 7GJ
>
> E-mail: [log in to unmask]
>
> http://scholar.google.co.uk/citations?user=ty6c5IMAAAAJ
>
> ________________________________________
> From: Stable Isotope Geochemistry [[log in to unmask]] On Behalf Of Rheane DA Silva [[log in to unmask]]
> Sent: 24 February 2016 11:17
> To: [log in to unmask]
> Subject: [ISOGEOCHEM] need help to improve standard deviation for H2 stable isotope analysis.
>
> Hello Colleagues,
>
> I would really appreciate your advice on the issue of analyzing stable H2 isotopes.
> I work primarily on n-alkanes and hence as an instrument set up step I ran 10 replicate runs of a C10-C40 n-alkane mixture. variable concentrations led to variable standard deviation values.
>
> For example C-23, a 25ppm standard showed 20.8 standard deviation and a 12ppm standard gave a standard deviation value of 1.9.
> in the case of C-18, the 25ppm gave a value of 3.05 and the 12ppm showed 14.7. similar results were reflected in a 90ppm n-alkane mixture, where the higher n-alkanes gave the worst standard deviation values.
>
>
> the GC parameters are as follows, rate Temp hold time
> 50 1.00
> 10 140 0.00
> 6.0 320 15.00
> PTV injector via an autosampler at 50 oC going up-to 320 at 3 oC/sec and flow rate 1.22 ml/min.
>
> the GCC is set at 1450 oC
>
> IRMS: 3 reference pulses for 20 secs each and back-flush switching off at 430 secs and switching on at 3100secs.
> the standard deviation for std-on-off was 0.4 and H3 factor was found to be 5.948962.
>
> I would really be grateful for any suggestions that can lead up to improving and stabilizing the standard deviation for n-alkanes
>
>
>
> Rheane Da Silva
> Senior Research Fellow(VNJCT)
> Gas Hydrates Research Group
> CSIR - National Institute of Oceanography
> Goa-India
> 403004
>
> ----------
>
> This email has been scanned for spam and viruses by Proofpoint Essentials cloud email security - visit the following URL to report this email as spam:
> https://eu1.proofpointessentials.com/index01.php?mod_id=11&mod_option=logitem&mail_id=56n05lWV%2BilU&rid=22280648&report=1
>
> Robert Gordon University is the top university for graduate jobs in the UK HESA July 2015
>
> Robert Gordon University, a Scottish charity registered under charity number SC 013781.
>
> This e-mail and any attachment is for authorised use by the intended recipient(s) only. It may contain proprietary material, confidential information and/or be subject to legal privilege. It should not be copied, disclosed to, retained or used by, any other party. If you are not an intended recipient then please promptly delete this e-mail and any attachment and all copies and inform the sender. Please note that any views or opinions presented in this email are solely those of the author and do not necessarily represent those of Robert Gordon University. Thank you.
>
|