Hi Geoff
You might try to check full scan or m/z 32 & 40, when needle valve is opened and closed. And check the vacuum is normal or not under both modes. It will help you to isolate where the main problem comes from.
If machine works good, you may check the He supply to the IRMS system is ok or not. I mean that maybe you checked the He pressure is ~ 1 bar on conflo but actually He is not supplying well. I meet similar situation before, and the problem is that there is not enough He going to the MS. It might comes from needle valve leaking, capillary(one or both) blocked/moved/broken.
Cheers,
NJ
Dr. Nai-Jung, Wan 萬乃容
H&N- Health & Nutrition健康暨營養事業群
3F,No. 125, Wu Kung Road, Wuku Industrial Zone, New Taipei City, Taiwan 24886
24886 新北市五股區(新北產業園區)五工路125號3樓
Email: [log in to unmask]
Phone: +886 2 2299 3279 Ext. 6732
-----Original Message-----
From: Stable Isotope Geochemistry <[log in to unmask]> On Behalf Of Geoff Schwaner
Sent: Tuesday, April 5, 2022 12:22 AM
To: [log in to unmask]
Subject: [EXTERNAL] [ISOGEOCHEM] Delta V Advantage Reference Dilution Issues
*** WARNING: this message is from an EXTERNAL SENDER. Please be cautious, particularly with links and attachments. *** ________________________________
Hello IsoGeoChem,
I am running a Delta V Advantage - EA Isolink CN - ConFlow IV setup, and have had an interesting issue come up. In the middle of a set of on/offs while in CO2 mode, the signal suddenly maxed out for all 3 cups. This was not a software bug or freeze up, the detector was simply maxed out. I turned off the Ref II main valve, and the issue still persisted. I had to turn off the CO2 valve as well before the signal started to very slowly decrease. I then checked the N2 configuration, which had similar issues. Mass 28 was maxed out, along with mass 30, and mass 29 was much higher than usual as well. Turning the N2 reference gas valves on or off does not seem to have any effect on the N2 signals, they stay that way regardless. Box and trap values seem normal, and it is a fairly new filament.
Troubleshooting I have already tried to resolve this:
Restarting the Delta V
Reseating the amplifier cards
Deleting and recreating dynexternal files Restoring to a previous working version handler file
When all of that failed to change anything, I started playing around with the tuning, and was able to get the signal intensities down by changing trap voltage, electron energy and extraction, but it is still not behaving correctly. One key thing I noticed: I tried re-determining reference gas dilutions, and the max dilution is ~1%. I tried running a few standards at increasing sample dilutions to see if sample dilution is still working, but the signal backgrounds are too messed up to detect peaks.
Could this be a He flow controller issue? Has anyone run into this issue before or think of any additional troubleshooting I can try?
Thank you,
Geoff Schwaner
Information in this email and any attachments is confidential and intended solely for the use of the individual(s) to whom it is addressed or otherwise directed. Please note that any views or opinions presented in this email are solely those of the author and do not necessarily represent those of the Company. Finally, the recipient should check this email and any attachments for the presence of viruses. The Company accepts no liability for any damage caused by any virus transmitted by this email. All SGS services are rendered in accordance with the applicable SGS conditions of service available on request and accessible at https://www.sgs.com/en/terms-and-conditions
|