| Sender: |
|
| Date: |
Thu, 18 Jul 2002 08:20:22 +0100 |
| Reply-To: |
|
| Content-type: |
text/plain; charset=US-ASCII |
| Subject: |
|
| From: |
|
| Content-transfer-encoding: |
7BIT |
| In-Reply-To: |
|
| Organization: |
Dundee University |
| MIME-Version: |
1.0 |
| Parts/Attachments: |
|
|
Dear Giovanni,
I'm afraid I beg to differ with Bruno's suggestion.
Since you wish to calibrate your lab's CO2 cylinder you have will to
enlist the help of either a professional service or a friendly
colleague with a dual-inlet machine (unless you've got one yourself).
If you wish to report your results against PDB or V-PDB, your CO2
has to be calibrated directly against an international standard on a
dual-inlet machine and not via a secondary standard using two
different instruments.
Bruno is right in saying there can be an off-set. The difference in
d13C values for e.g. a given FAME between EA-IRMS and GC-
IRMS measurements using the same reference gas cylinder, is a
result of the external isotopic calibration technique in GC-IRMS
when used in conjunction with a software that does not
compensate for the isotopic fraction of the CO2 peak from the on-
line combusted FAME ("time-shift correction").
There are ways around this problem but that's a different story.
Cheers,
Wolfram
> Dear Jovanni,
>
> your can buy a FAME, measure it with EA-IRMS (calibrated against NIST
> or IAEA certified standards) and then you can adjust you GCC-IRMS CO2
> (which will be different from the EA-IRMS value) by measuring the same
> FAME with GCC-IRMS Hope this helps
>
> Kind regards
>
> Bruno
>
> (Dr. Bruno Glaser)
> Institute of Soil Science and Soil Geography
> Bayreuth
> >
> Hi all,
> >
> > It's the first time I need to calibrate the CO2 (delta13C) of our
> > GCC-IRMS apparatus: has anybody some suggestion on how to do that?
> >
> > Thank you,
> > Giovanni Calderone
> >
> >
> > **************************************************************
> > EC - Joint Research Centre
> > e-mail: [log in to unmask]
|
|
|