Stefano
This error usually appears if the carrier gas pressure is set too low. Please
check that the outlet pressure on the He tank regulator is set to 300kPa and
the carrier pressure in the instrument is set to 160 kPa.
To clear this error from the display you must turn off the instrument for 10
seconds and then turn it back on again.
Cheers
Peter Stow
Stefano Bernasconi wrote:
> dear all
>
> We are experiencing a problem with our Carlo Erba NCS2500 elemental
> analyzer. When we start a run we get the message "error: gas flow out of
> range".
> We checked for the presence of leaks and the gas seems to flow correctly.
> Could it be a problem be caused by the sample carussel? Did anybody
> experience the same problem before?
>
> thank you
>
> stefano
>
> -------------------------------