Tue, 28 Sep 1999 12:33:15 +0100
|
> Dear Group,
> I have been investigating stable carbon isotope ratios of
> organic contaminants in the sub-surface of a disused manufactured gas
> plant. This involves injecting solvent extracts and getting the
> mass spectra and d13C ratio. Whenever molecular sulphur is identified on
> the quadropole, the 46/44 ratio for the background subtraction on the IRMS
> (Isoprime, Micromass) goes off the scale, making automated or manual
> correction practically impossible. Has anyone any ideas what is causing
> the 46 ion? I can't figure it out. I hope the answer is not too obvious!
> Thank you
Dear James,
I'm afraid it just might be. Obvious that is. The answer that is.
If you have a look at the EI mass spectrum of CS2 you will find
fragment ions at m/z44 and m/z46 in the predicted ratio for natural
abundant S isotopes. These fragments correspond to 12C32S and
12C34S, respectively.
You either have contamination of CS2 present or it is formed in-
situ. The EI mass spectrum of CS2 contains the following peaks in
decreasing order of intensity: m/z76, mol peak and base peak (and
its M+2), m/z32 (and 34) and m/z44 (and 46).
Hope this helps.
All the best,
Wolfram
*****************************************************