Hilary,
I now back flush the GC column after every sample to get rid of these
strange trace interfering peaks for CO analysis. I use a 2m by what looks
to be 6 mm OD molecular sieve 5A column from Costech Analytical
Technologies. It is a rather large diameter but this diameter has to be
used for the flow rate of 100 mL min-1 I am using. It separates N2 and CO
fine at 64 degrees C GC oven temperature. Although the CO peak tails more
than I would like, the precision when injecting CO through the GC column is
plus or minus 0.06 delta units, which is not significantly different than I
get with ref gas injection, so apparently the trailing does not affect
precision significantly.
Paul.
At 02:01 PM 2/24/04 +1100, you wrote:
>Thanks Toti
>
>Could I have a copy of the application notes?
>
>Now there's a good question about the CO2!
>
>Yes, it does come off, as rather a blurred peak at 1000 seconds under our
>conditions. So, if you do a 600 second run, your CO2 comes off RIGHT in the
>tail of your CO. And although the amplitude is very small after scrubbing,
>the dissociation in the source causes some wild values. But even without
>the CO2 -> CO+O fractionation, the gas comes from the PREVIOUS sample.
>
>So you have to scrub VERY well or do long runs. And the CO2 doesn't have
>enough amplitude (from our set up) to show on anything apart from the ratio
>traces. And then only for about 1 in 20 samples ...
>
>Hil
>
>.
Paul D. Brooks,
Center for Stable Isotope Biogeochemistry,
Dept. Integrative Biology MC3140,
3060 Valley Life Sciences Building,
UC Berkeley, Ca. 94720-3140.
[log in to unmask]
phone (510)643-1748,
FAX (510)643-1749