Thu, 24 Jul 2008 14:15:30 -0700
|
Marc,
> I see vast nonlinearity when analyzing a water sample. Reference
>and sample peaks look fine and sample peak intensities are as
>expected for the sample weight. It seems like, in contrast to the
>reference gas ratio calculations, the H3 factor is not included in
>the ratio calculations for the samples. The technician and I have no
>more ideas. It is the first time D/H is measured on this machine.
>
>First of all: Is the H3 correction for the samples commonly
>calculated automatically by the mass spec software or do you have to
>correct the sample ratios manually for the H3 contribution?
>(Perhaps there is someone using the same isoprime mass spec and
>could tell me about some hidden places in the Ion Vantage software
>where a simple tick could solve all my problems...??)
If the Ion Vantage software is similar to the MassLynx software used
to run an Isoprime, there is indeed a box that needs a tick for the
software to use the H3+ correction factor when calculating a final
delta value.
In MassLynx, you go to Edit Species, select H2, choose Reference Gas,
and then the Molecule tab. In the H3+ Correction section, there's a
box labelled "Enable" which needs to be checked for the calculations
to use the H3+ correction factor. Incidentally, the "Use H3+
calibration data" box must have a check for the software to
re-calculate the H3+ correction factor when you perform the routine
to measure the H3+ correction factor. I haven't used or seen the Ion
Vantage software, but I imagine that it has something similar.
I hope this helps, and good luck!
Cheers,
Simon
--
*****
Simon Poulson
Research Associate Professor
Dept. of Geological Sciences & Engineering MS-172
University of Nevada - Reno
1664 N. Virginia St.
Reno, NV 89557-0138. USA
(775) 784-1104
Fax: (775) 784-1833
[log in to unmask]
http://www.mines.unr.edu/isotope/simonhome.html
*****
|
|
|