Skip Navigational Links
LISTSERV email list manager
LISTSERV - LIST.UVM.EDU
LISTSERV Menu
Log In
Log In
LISTSERV 17.5 Help - ISOGEOCHEM Archives
LISTSERV Archives
LISTSERV Archives
Search Archives
Search Archives
Register
Register
Log In
Log In

ISOGEOCHEM Archives

Stable Isotope Geochemistry

ISOGEOCHEM@LIST.UVM.EDU

Menu
LISTSERV Archives LISTSERV Archives
ISOGEOCHEM Home ISOGEOCHEM Home

Log In Log In
Register Register

Subscribe or Unsubscribe Subscribe or Unsubscribe

Search Archives Search Archives
Options: Use Classic View

Use Monospaced Font
Show HTML Part by Default
Show All Mail Headers

Topic: [<< First] [< Prev] [Next >] [Last >>]

Print Reply
Re: GC-C-IRMS issue
River He <[log in to unmask]>
Tue, 13 Oct 2009 17:52:41 -0700
text/plain (465 lines)
Hi Berb,

We do change the liner and septa rountinely but not as often as you guys do; we change septum every two days and liner every week. But certainly I can try to do this every day to see how it works.  I used to do bake the column at the end of each run but stopped doing that for some reasons. Our problem has been there since this instrument was bought two years and certainly before I came over.

River

--- On Tue, 10/13/09, Herbert Tobias <[log in to unmask]> wrote:

> From: Herbert Tobias <[log in to unmask]>
> Subject: Re: [ISOGEOCHEM] GC-C-IRMS issue
> To: [log in to unmask]
> Received: Tuesday, October 13, 2009, 8:27 PM
> River,
> 
> We run underivatized steroids at Cornell University as
> well.  My experience is that if the column seems to work
> initially, it's the inlet liner and septum.  You need
> to change the liner and septum every day.  The other thing
> to do is make sure you have at least a few minutes of the
> temperature program of each run at the highest bakeout
> temperature.  If you don't continually bakeout the
> column, you'll get the problems you describe.
> 
> 
> cheers,
> -Herb
> 
> On Tue, Oct 13, 2009 at 8:09 PM,
> River He <[log in to unmask]>
> wrote:
> 
> Dr. Meier-Augenstein,
> 
> 
> 
> I forgot to mention that we used to use guard column at the
> front end of GC column.  But it seems to us that this
> doesn't really protect the GC column, as we still see
> the similar chromatograph problem with such guard column.
> 
> 
> 
> 
> River
> 
> 
> 
> --- On Tue, 10/13/09, Wolfram Meier-Augenstein <[log in to unmask]>
> wrote:
> 
> 
> 
> > From: Wolfram Meier-Augenstein <[log in to unmask]>
> 
> > Subject: Re: [ISOGEOCHEM]
> GC-C-IRMS issue
> 
> > To: [log in to unmask]
> 
> > Received: Tuesday, October 13, 2009, 2:55 AM
> 
> > Dear River He,
> 
> >
> 
> >
> 
> > Do you run the steroids underiviatized or derivatized?
> If
> 
> > yes to the
> 
> > latter, what derivatisation do you use?
> 
> >
> 
> > Also, do you use a retention gap (= 1 m deactived FS
> 
> > capillary between
> 
> > injector and column)?
> 
> >
> 
> > How do you inject; split or splitless?
> 
> >
> 
> > Lastly, have you checked your column/s after initial
> 
> > conditioning for
> 
> > active sites by running a Grob test? Similarly, do
> you
> 
> > check yor
> 
> > column/s one the problem starts to manifest itself
> with a
> 
> > Grob test run?
> 
> > An alkane test will tell you next to nothing if a
> column
> 
> > has developed
> 
> > active sites since active sites predominantly affect
> only
> 
> > polar
> 
> > compounds.
> 
> >
> 
> >
> 
> > Best,
> 
> >
> 
> > Wolfram
> 
> >
> 
> >
> 
> > ****************************
> 
> > Dr W Meier-Augenstein, CChem, FRSC
> 
> >
> 
> > Extension: 2624
> 
> > E-mail: [log in to unmask]
> 
> > URL: http://www.scri.ac.uk/staff/wolframmeieraugenstein
> 
> >
> 
> > -----Original Message-----
> 
> > From: Stable Isotope Geochemistry [mailto:[log in to unmask]]
> 
> > On
> 
> > Behalf Of River He
> 
> > Sent: 13 October 2009 07:45
> 
> > To: [log in to unmask]
> 
> > Subject: [ISOGEOCHEM] GC-C-IRMS issue
> 
> >
> 
> > Dear All,
> 
> >
> 
> > I am posing this note to ask any suggestion and
> advice
> 
> > about the
> 
> > problems we have been struggling with our GC-C-IRMS
> for
> 
> > quite a while.
> 
> > In fact, we are a little confused. 
> 
> >
> 
> > In our lab, we have a Thermo GC-C-IRMS (Delta V Plus)
> for
> 
> > d13C analysis
> 
> > of steroids. Since we bought this system, we have
> begun to
> 
> > deal with the
> 
> > chromatograph problem; the tailing of sample peaks. 
> 
> > Normally after
> 
> > putting on a new GC column, the first few batches are
> fine,
> 
> > especially
> 
> > the first one or two. BUT very quickly, the
> chromatograph
> 
> > of samples
> 
> > deteriorates; the tailing gradually shows up and gets
> worse
> 
> > and worse at
> 
> > very fast rate, and after few weeks, the tailing is so
> bad
> 
> > that we
> 
> > cannot analyze any sample without changing a new
> column.
> 
> >
> 
> > System ontamination by sample was easily considered a
> 
> > primary cause, as
> 
> > once a while we do see "dirty" samples with
> high CO2
> 
> > background. Cutting
> 
> > column, cleaning injector and changing liner or septa
> 
> > barely work for
> 
> > us(well only worked for once or twice probably).  In
> 
> > fact, we do
> 
> > routinely cut column, change septa, liner and
> condition
> 
> > column. Then, we
> 
> > also thought the column may decay due to leaking or
> 
> > moisture from the
> 
> > samples. Keeping this in mind, in the last few months,
> we
> 
> > had closely
> 
> > monitored these two factors and it seems to us that
> they
> 
> > are not the
> 
> > reasons either. The Ar background of our system at
> straight
> 
> > mode is
> 
> > alway at the level of 80 mV with oven temperature at
> 60 C
> 
> > and its
> 
> > reading only increases 20 to 30 mV at the temperature
> of
> 
> > 310 C.
> 
> >
> 
> > Another thing we observed also kind of confuses us
> more.
> 
> > Although the
> 
> > chromatographs of our steroids samples are very poor
> with
> 
> > huge tailing,
> 
> > but we can get pretty decent peaks when we run mixture
> of
> 
> > alkanes. This
> 
> > tells us that the column seems okay.  I am not sure
> 
> > whether the
> 
> > oxidation tube is a problem.  The current one is
> 
> > relatively new and we
> 
> > just changed it in August.  Also, we regenerate
> 
> > oxidation tube for at
> 
> > least one hour between batches and do this at least
> 
> > overnight during
> 
> > weekend. 
> 
> >
> 
> > We will greatly appreciate if anyone can give us
> suggestion
> 
> > and advice
> 
> > to cope with our problems.  BTW, the GC colum
> 
> > wecurrently use is
> 
> > DB-17MS.
> 
> >
> 
> > THanks in advnace,
> 
> >
> 
> > River He
> 
> > SMRTL, USA
> 
> >
> 
> >
> 
> >      
> 
> >
> __________________________________________________________________
> 
> > Yahoo! Canada Toolbar: Search from anywhere on the
> web, and
> 
> > bookmark
> 
> > your favourite sites. Download it now http://ca.toolbar.yahoo.com.
> 
> >
> 
> >
> ______________________________________________________________________
> 
> > This email has been scanned by the MessageLabs Email
> 
> > Security System.
> 
> > For more information please visit http://www.messagelabs.com/email
> 
> >
> ______________________________________________________________________
> 
> >
> 
> >
> ______________________________________________________
> 
> > SCRI, Invergowrie, Dundee, DD2 5DA. 
> 
> > The Scottish Crop Research Institute is a charitable
> 
> > company limited by guarantee.
> 
> > Registered in Scotland No: SC 29367.
> 
> > Recognised by the Inland Revenue as a Scottish Charity
> No:
> 
> > SC 006662.
> 
> >
> 
> >
> 
> > DISCLAIMER:
> 
> >
> 
> > This email is from the Scottish Crop Research
> Institute,
> 
> > but the views expressed by the sender are not
> necessarily
> 
> > the views of SCRI and its subsidiaries.  This email
> and
> 
> > any files transmitted with it are confidential to the
> 
> > intended recipient at the e-mail address to which it
> has
> 
> > been addressed.  It may not be disclosed or used by
> any
> 
> > other than that
> 
> > addressee.
> 
> > If you are not the intended recipient you are
> requested to
> 
> > preserve this confidentiality and you must not use,
> 
> > disclose, copy, print or rely on this e-mail in any
> way.
> 
> > Please notify [log in to unmask]
> 
> > quoting the name of the sender and delete the email
> from
> 
> > your system.
> 
> >
> 
> > Although SCRI has taken reasonable precautions to
> ensure no
> 
> > viruses are present in this email, neither the
> Institute nor
> 
> > the sender accepts any responsibility for any viruses,
> and
> 
> > it is your responsibility to scan the email and the
> 
> > attachments (if any).
> 
> >
> ______________________________________________________
> 
> >
> 
> 
> 
> 
> 
>     
>  __________________________________________________________________
> 
> Make your browsing faster, safer, and easier
> with the new Internet Explorer® 8. Optimized for Yahoo! Get
> it Now for Free! at http://downloads.yahoo.com/ca/internetexplorer/
> 
> 
> 
> 
> 
> -- 
> Herbert J. Tobias
> Cornell University
> Ithaca NY 14853
> 1-607-255-9009 (Lab)
> 
> 


      __________________________________________________________________
Looking for the perfect gift? Give the gift of Flickr! 

http://www.flickr.com/gift/

ATOM RSS1 RSS2

LIST.UVM.EDU CataList Email List Search Powered by LISTSERV