German,
Check the cup configuration switcher behind top front panel. If recently that
panel was moved you could accidentally switch system onto more sensitive
resistor.
Sergey.
>===== Original Message From Stable Isotope Geochemistry
<[log in to unmask]> =====
>We have a problem with our Conflo-III system, which is attached to a Delta
>Plus XL. When we turn on any of the reference gases (CO2 or N2), we get a
>large signal (>14 volts). My first thought was that the pressure was too
>high. We decreased the pressure using the gauge in the Conflo unit, but
>there was no luck. We checked the pistons and the distances of the
>capillaries inside the open split glass cell when they are on and off, and
>all of them are ok. I also thought that we were not getting any or low flow
>of helium. We measured the flow of the helium stream entering the open split
>glass cell and it is what is expected (~2 ml/min). We even disconnected the
>EA to see if somehow the problem was with the EA, but we still see the same
>problem. What makes this problem so particular is that when we turn on the
>CO2 reference gas we only get a large signal on mass 44. Masses 28, 40, or
>18 only show a slight increase. When we turn on the N2 reference gas we only
>get a large signal (>14 volts) on mass 28, not on any other mass. I ran out
>of ideas, so I would really appreciate any help.
>
>Thanks,
>
>German