Skip Navigational Links
LISTSERV email list manager
LISTSERV - LIST.UVM.EDU
LISTSERV Menu
Log In
Log In
LISTSERV 17.5 Help - ISOGEOCHEM Archives
LISTSERV Archives
LISTSERV Archives
Search Archives
Search Archives
Register
Register
Log In
Log In

ISOGEOCHEM Archives

Stable Isotope Geochemistry

ISOGEOCHEM@LIST.UVM.EDU

Menu
LISTSERV Archives LISTSERV Archives
ISOGEOCHEM Home ISOGEOCHEM Home

Log In Log In
Register Register

Subscribe or Unsubscribe Subscribe or Unsubscribe

Search Archives Search Archives
Options: Use Forum View

Use Proportional Font
Show Text Part by Default
Show All Mail Headers

Message: [<< First] [< Prev] [Next >] [Last >>]
Topic: [<< First] [< Prev] [Next >] [Last >>]
Author: [<< First] [< Prev] [Next >] [Last >>]

Print Reply
Subject:
Re: help for tripping source and pumps - MAT252
From:
John Hill <[log in to unmask]>
Reply To:
Stable Isotope Geochemistry <[log in to unmask]>
Date:
Thu, 4 Dec 2008 14:50:56 +0000
Content-Type:
text/plain
Parts/Attachments:
text/plain (96 lines)
Hi Alison

Sounds to me that you could be getting a discharge from the HT inside 
the ion source which is causing the vacuum to trip. This is due to a 
transient spike going round the earth line. The discharge could be 
caused by the connecting wires inside the source perhaps very close to 
the grounded heater wires or possibly a crack in the ceramic tube that 
connects the inlet line to the ion source itself. You will need to take 
the source out and thoroughly check all the connections.

Regards

John Hill
Mass Spectrometry Facility
Chemistry Dept.
University College London

Alison Pye wrote:
> I’m hoping someone can help me with a problem that has developed on my 
> MAT252. I recently went through a flurry of bad refurbished filaments 
> (no trap emission) and now that I have a nice new filament in there, I 
> can’t get the source, or pumps, to stay on more than a couple minutes 
> after turning the voltage up to the full 10 KV. I haven’t yet opened the 
> ConFlo, so it is in dual inlet mode. I left it overnight with the HV at 
> 8 KV, and the pumps were still going this morning, but the HV was off 
> (source control still on). HV reset OK. Vacuum was at 3.6 E-9. Within 2 
> minutes of turning on the HV to 10 KV, the pumps tripped again. I’m 
> lucky now in that after half a dozen trips over the last three days, 
> I’ve at least figured out how to reset the pumps before they vent so 
> that I don’t lose vacuum. As long as I am in the room.
> 
>  
> 
> I’ve had a similar problem to this in the past, and the problem was with 
> a couple capacitors the vacuum gauge monitor, which was giving faulty 
> signals causing the trips. At that time, the needle valve on the 
> analogue vacuum display had been “jittery” and then outright failed. And 
> the inlet forevacuum display in Isodat had been jumping around. But I 
> haven’t seen the same symptoms this time, except perhaps that the vacuum 
> during CF mode seemed slightly lower than normal (2.6 E-6 as opposed to 
> 4 E-6). I note that the vacuum gauge display is not responding at all 
> when the pumps trip, until I can reset the MS, restart Isodat system, 
> and toggle the vacuum display buttons. Toggling the buttons is a trick 
> that has always been necessary from time to time to get the analogue 
> display to respond. Until the analogue vacuum gauge display returns to 
> an active reading mode, the pumps will not restart. As with my earlier 
> problems, this suggests a possible problem with the vacuum gauge monitor 
> again. However, I don’t recall there being a link before to having this 
> happen only with HV at the full 10KV.
> 
>  
> 
> And here is another piece to the puzzle… I just decided to keep an eye 
> on the source forevacuum (through the analogue display since I can’t 
> figure out how to get it to display through Isodat) to see if it was 
> displaying any odd behaviour. Since I have switched it to this setting, 
> still at 10 KV, it hasn’t tripped yet. As I write this it’s only been 1 
> ½ hours, but it hasn’t tripped yet. I couldn’t get more than 2 minutes 
> of run time this morning and yesterday when set at 10KV while monitoring 
> source high vacuum.
> 
>  
> 
> Does anyone out there recognize these symptoms? At this point I’m still 
> thinking it is the vacuum gauge monitor. I should know by the end of the 
> day whether this is directly related to the analogue display being set 
> to the source high vacuum at 10 KV only, or whether it is a fluke that 
> it hasn’t tripped again.
> 
>  
> 
> Thanks for any insight.
> 
>  
> 
> --Alison
> 
>  
> 
> ------------------------------------------------------------------------
> 
> Alison Pye  •  [log in to unmask] <mailto:[log in to unmask]>
> Stable Isotope Lab Coordinator
> 
> CREAIT • TERRA Facility • Stable Isotope Lab 
> <http://www.mun.ca/creait/TERRA/SIL.php>
> Alexander Murray Building
> Memorial University of Newfoundland
> St. John's, NL • Canada • A1B 3X5
> ph: (709) 737-3217 • fax: (709) 737-2589
> 
>   
> 
>  
> 

ATOM RSS1 RSS2

LIST.UVM.EDU CataList Email List Search Powered by LISTSERV