Hi Charlotte,
If your purifier is the heated type, restoring power to it is going to
release a bunch of junk downstream. Just something to think about if
the system was ever open to it before you did this calibration.
Also, the slope in my 13C normalizations has ranged from 0.94 to 1.05
depending on the run so I don't know if your slope estimate is really
that bad. I'd make sure you have multiple points spanning the broadest
range of deltas, all at similar amplitude.
For EA work, I haven't noticed any problems in std on offs or precision
of replicates when water is around 3V.
How is the precision of your CO2 on-offs in this conflo and any other
reference splits? I wonder how clean your tank is.
Cheers,
Ben Harlow
Washington State University
Stable Isotope Core Laboratory
-----Original Message-----
From: Stable Isotope Geochemistry [mailto:[log in to unmask]] On
Behalf Of Charlotte Lehmann
Sent: Monday, February 05, 2007 7:53 AM
To: [log in to unmask]
Subject: [ISOGEOCHEM] slope of standard curve for d13C
I recently ran NIST standards to develop a new standard curve for d13C
bulk isotopes on our Delta V and am wondering to what to attribute the
fact that the slope is less than 1.0. This time it was 0.93, resulting
in enriched values that are much more enriched than they were six
months ago. We have a new tank of Scientific CO2 for the reference
gas, and it appears to have a d13C value of about -39.4 per mil
(whereas our previous tank from the same supplier was -22 per mil).
Any suggestions?
Also, and I don't know if this is contributing to the above, we have a
fair amount of water hanging around as evidenced by Mass 18 values in
the several thousand mV range, with no indication of leaks (Mass 40 is
low). I just discovered that the gas purifier on the GC-GCC III side
had a blown fuse. But the problem is evident on the EA-Conflo III side
of things and we have never used a gas purifier there and not had this
problem before to my recollection. Heaters are on at the valves and
the source. We already replaced the needle valves with Nupro valves.
Spent some time running diagnostics last week: focus is good;
linearity checks out. Amplifier test passes (although I can't pick the
gas configuration to use because it only sees "CO" which we do not
routinely use. Peak shape and flatness are good. System stability is
excellent. Signal stability (on CO2) seems to be 2.5 times what the
manual says it should be; not sure what that means or how to correct
it...
I'll be interested to hear any ideas!
--