Hi Million,
First thing to check would be for continuity across the filament, just
in case you've got a DOA filament (it happens).
As you have mentioned yourself already, this could just be a slight
misalignment of the source leading to an eclipsed alignment of the gas
inlet like so (), or worse. Runnung a magnet background scan first is
definitely a good idea. Unless there is something seriously wrong with
the source (no emission) or the analyzer, you should see peaks for N2
background at around 8,500 steps, give or take.
If you see the N2 peaks suggesting there is nothing wrong with either
source or analyzer per se, switch the source off, open the door and
disconnect all the supply cables.
If you are careful you can get away with not turning the pumps off but
this is up to you. Loosen the bolts ever so slightly until you can
wiggle the source flange left to right and back. Twist it to the left
(gently does it) and mark the stop position with a felt pen (across
housing and flange at the 9 o'clock position where the alignment marker
is). Now twist it to righ until it will go no further and mark this
position too. Twist to the left again until you are halfway between the
either extreme. This should now have the gas inlet holes aligned like
so: O.
Best,
Wolfram
****************************
Dr W Meier-Augenstein, CChem, FRSC
Extension: 2624
E-mail: [log in to unmask]
URL (SCRI): http://www.scri.ac.uk/staff/wolframmeieraugenstein
URL (MRS): http://www.isotopesignatures.co.uk
-----Original Message-----
From: Stable Isotope Geochemistry [mailto:[log in to unmask]] On
Behalf Of Million Hailemichael
Sent: 11 July 2010 21:07
To: [log in to unmask]
Subject: [ISOGEOCHEM] no ion beams after source cleaning
I removed and cleaned the ion source for our delta V advantage. I
changed a new filament and I put back the source and everything looks
fine on isodat screen. Values are...
HV= 3.01KV
Box= 0.72 mA
Trap= 0.77 mA
vacuum= 1.4e-006
Magnet (steps)= 11572
When I select reference gas (CO2) on the conflo screen to do a jump
calibration no CO2 gas is flowing into the system. I checked for a leak
and NO leak, we have a half tank CO2 gas. It seems ions are not reaching
the cups or ions are not generated. I checked the CO2 tank is open,
conflo is open and SGE valve is open.
I am planning to remove the source again and see if the small opening in
the side of the ion source chamber is somehow blocked or miss aligned
(even though I checked every thing before I put it back) . Or when I
pushed the source into the mass spectrometer possibly a wire may have
been disconnected?.
Any advice or suggestions before I remove the ion source again to check
for any loose connection.
Thanks
Million
______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email
______________________________________________________________________
______________________________________________________
SCRI, Invergowrie, Dundee, DD2 5DA.
The Scottish Crop Research Institute is a charitable company limited by guarantee.
Registered in Scotland No: SC 29367.
Recognised by the Inland Revenue as a Scottish Charity No: SC 006662.
DISCLAIMER:
This email is from the Scottish Crop Research Institute, but the views expressed by the sender are not necessarily the views of SCRI and its subsidiaries. This email and any files transmitted with it are confidential to the intended recipient at the e-mail address to which it has been addressed. It may not be disclosed or used by any other than that addressee.
If you are not the intended recipient you are requested to preserve this confidentiality and you must not use, disclose, copy, print or rely on this e-mail in any way. Please notify [log in to unmask] quoting the name of the sender and delete the email from your system.
Although SCRI has taken reasonable precautions to ensure no viruses are present in this email, neither the Institute nor the sender accepts any responsibility for any viruses, and it is your responsibility to scan the email and the attachments (if any).
______________________________________________________
|