HI Paul--
They have the safety switch on the door that gives you access to the
source since opening that gives you access to exposed high voltage
lines. It's fine to open that up with power off to check for shorts
across the different plates as well to box/trap/ground. But if you were
to defeat the safety switch and check the voltages on the board attached
to the source, that would be dangerous. Even if you dropped voltage to
1000 V. Not recommending this. You don't want to accidentally short
things to each other or to ground.
take care,
gerry
p.s. good luck and be very very careful...
p.p.s. Those resistors are set up to minimize damage if there's a bad
short, pg 7.4 of catalog--maybe check them to see if they're ok with
power down and disconnected before testing anything live.
On 1/22/2016 2:00 PM, Paul Eby wrote:
> Andy,
>
> Thank you for your response. I'm not sure, though, that we have the same board.
>
> The trap on my board (small one attached to the outside of the source feed through, PN2077590-00) has the trap attached to pin A2 (GND is A13, box is A4). I don't see any pins labelled in the format you describe....
>
> Paul
>
> -----Original Message-----
> From: Stable Isotope Geochemistry [mailto:[log in to unmask]] On Behalf Of Andrew Ross
> Sent: January-22-16 9:36 AM
> To: [log in to unmask]
> Subject: Re: [ISOGEOCHEM] Delta source measurements
>
> Hi Paul,
>
> On our Delta-V Plus (same source control board as the advantage but w/o the quad control - pretty sure), we measure across pins MP313 (common) and MP317 for the trap voltage measurement (usually 0-40V). Electron Energy is J348 (vs MP313).
>
> Good Luck,
> Andy Ross (CEOAS, OregonStateU)
>
> -----Original Message-----
> From: Stable Isotope Geochemistry [mailto:[log in to unmask]] On Behalf Of Paul Eby
> Sent: Friday, January 22, 2016 9:22 AM
> To: [log in to unmask]
> Subject: [ISOGEOCHEM] Delta source measurements
>
> Everyone,
>
> My Delta V Advantage is acting oddly, and I'd like to measure the trap voltage to verify that it is working properly. Does anyone know which pin on the board I would have to measure against? Not the board ground, which is offset 3kV.
>
> - Yes, I know that the board, when live, carries 3kV and is dangerous
> - Yes, I know that measuring vs the wrong points will blow sh!t up
> - Yes, I know the cabinet has a lockout
>
> I've done this on other Delta versions, but don't have schematics or details for the board on this Delta V Advantage.
>
> Paul Eby
>
>
> Water Management
> Alberta Innovates - Technology Futures
> Vancouver Island Technology Park
> 3 - 4476 Markham St., Victoria, BC, Canada V8Z 7X8
>
> Phone: 250-483-3290 | Fax: 250-483-1989
> [log in to unmask]
> AlbertaTechFutures.ca
>
|