Dear All:
We changed the heater of trap 2 only once in the past six years. Trap 1
is the problematic one. We change it in 6 to 12 months till we have
problem recently.
Best.
Dachun
On 2016-05-13 12:06, Matthew Lachniet wrote:
> Hi all,
>
> How long do your heater cartridges and thermocouples routinely last?
>
> I usually have to replace every 6-12 months with no use of compound. I
> used the fan for several years but it never seemed to make a
> difference so I removed it when it failed.
>
> Fortunately it is dry in the Mojave Desert so ice and water isn't so
> much an issue.
>
> Matt
>
> Matthew Lachniet
> Professor of Geoscience
> University of Nevada Las Vegas
>
> -----Original Message-----
> From: Stable Isotope Geochemistry [mailto:[log in to unmask]] On
> Behalf Of David A. Mucciarone
> Sent: Friday, May 13, 2016 11:58 AM
> To: [log in to unmask]
> Subject: Re: [ISOGEOCHEM] Frequently broken heater and thermocouple of
> Kiel
>
> The thermal joint compound I use is from Type 120, Wakefield
> Engineering, Beverly, MA 01915, 781-406-3000. I did not Google it,
> hopefully you can do that part. For what it is worth, I tried just
> using the compound (no fan) only without using the foil and the
> cartridge failed early, better than without, but early nonetheless.
>
> Good luck,
>
> Dave
>
> On 5/12/2016 3:43 PM, zhangd wrote:
>> Dear David, Andy and others:
>>
>> Thanks you for the suggestions. I am going to use fan and the
>> conducting compound. What brand of that compound you use, David?
>>
>> Thanks.
>>
>> Dachun
>>
>>
>>
>> On 2016-05-12 14:47, David Mucciarone wrote:
>>> I'm not sure what KC you have so this may or may not be helpful.
>>> Depending on how the previous heat cartridges came out it is possible
>>> that during removal the hole enlarged. If this happened, water can
>>> work it's way in and cause early failure. This happened on my KCIII.
>>> The solution was to shim with copper soil (had it on hand, other foil
>>> could be used), given the thickness of my foil, I simply wrapped the
>>> cartridge with one layer of foil to provide a snug fit. I then used a
>>> conducting compound to seal the top. This is a non hardening
>>> compound, I am on the road, but will be in the lab, hopefully on
>>> Friday (if my flight ever gets off the ground) in the event you need
>>> more info on this potential solution.
>>>
>>> David
>>>
>>> Sent via the Samsung GALAXY S®4, an AT&T 4G LTE smartphone
>>>
>>> -------- Original message --------
>>> From: zhangd <[log in to unmask]>
>>> Date: 05/12/2016 14:55 (GMT-05:00)
>>> To: [log in to unmask]
>>> Subject: Frequently broken heater and thermocouple of Kiel
>>>
>>> Dear Colleagues:
>>>
>>> Both the new heating cartridge and measuring resistor in our Kiel
>>> device trap 1 are broken (not the same time) in less than two month.
>>> It happened twice already. Any suggestions would be appreciated.
>>>
>>> Thanks.
>>>
>>> Dachun Zhang
>>> ESS dept UCI
>> .
>>
>
>
> --
> Mailing address:
> David A. Mucciarone
> 473 Via Ortega, Rm. 140
> Stanford, CA 94305
> 650-723-0817(O), 650-804-6403(M)
>
> Ship to address:
> David A. Mucciarone
> 367 Panama Street, Rm. 50
> Stanford, CA 94305
|