HI All--

Yeah, NOx's tend to be "sticky" and not pump out quickly.

Something a little different--it might be getting time to change the oil 
(oil pads) on the turbo pumps--especially if water is high in the mass 
spec.  And double check the roughing pumps too--if the rotor seal leaks, 
you can still run it for a while, but with increased backstreaming.

It doesn't sound like you can get away with conditioning the source, and 
I'd guess the source is fairly clean.

take care,

gerry

On 9/12/2019 9:23 AM, Arnold,Thomas E wrote:
> As a follow up -
>
> I baked out the EA oven, closed the valve to the IRMS and baked that out for 24hrs. We still have the same issue with mass 30 increasing within a single on/off run and across multiple on/off runs. Today I will switch out our He tank for a new one, and change the water trap (even though it show no sign on water consumption).
>
> Not sure if there are any other suggestions beyond what has already been put forth. The standard deviations look fine after the 4th (or so) N2 on/off run, so maybe I am making too much of an issue out of this, but it is frustrating that such an issue would suddenly appear and be difficult to fix.
>
> Thanks for the help,
> -TEA
>
>> On Sep 10, 2019, at 2:08 PM, Harlow, Benjamin <[log in to unmask]> wrote:
>>
>>
>> EXTERNAL EMAIL: Exercise caution with links and attachments.
>>
>>
>>
>>
>> Tom, it works nice to install a 3 way valve at the exit of the EA. That way you can quickly vent to the room (or a flow meter) instead of the conflo. You can also use this valve to block off all flow out of the EA for gas savings or a leak down test.
>>
>> I would turn off the HF on your conflo and vent the EA whenever you do maintenance such as chemical changes and GC bake out. Then restore the EA to work mode and let it get up to flow and temperature for an hour or so before switching the 3 way back to the conflo. This allows any contaminants introduced during the maintenance to bake off and flow somewhere other than the IRMS.
>>
>> Have your helium or O2 tanks been changed recently?
>>
>> Cheers, Ben
>>
>> -----Original Message-----
>> From: Stable Isotope Geochemistry <[log in to unmask]> On Behalf Of Arnold,Thomas E
>> Sent: Tuesday, September 10, 2019 10:33 AM
>> To: [log in to unmask]
>> Subject: Re: [ISOGEOCHEM] Mass 30 instability on Delta V
>>
>> Hi Ben,
>>
>> We have been using the same UHP tank of N2 for some time now (predating the issue at hand).
>>
>> I have not baked the GC oven in the EA before. When you say vent the EA away from the conflo, can I do this by simply turning off (in insodat) the high flow capillary that is going into the conflo? If I remove the line in front of our Costech EA that is carrying the gas from the machine to the conflo, the N2 backgrounds skyrocket (as expected). Thus, I’m not sure which is the best way to vent it away from the conflo.
>>
>> Thanks,
>> TEA
>>
>>
>>> On Sep 10, 2019, at 12:37 PM, Harlow, Benjamin <[log in to unmask]> wrote:
>>>
>>>
>>> EXTERNAL EMAIL: Exercise caution with links and attachments.
>>>
>>>
>>>
>>>
>>> Hi Tom,
>>>
>>> What are your backgrounds? You could have a bad tank as well. Have any been changed recently? If 30 is increasing throughout your EA sequences it could be that the water scrubbing chemicals are not good, or there is a pathway through the trap. When was the last time the GC was baked? (Do this with the EA vented away from conflo)
>>>
>>> What I meant about instability is that is usually a consequence of excessive water. If you have good N2 on-off stability then I would think the water in your system is manageable. Usually, when we see a strong stair step effect in m/z 30 N2 is not reproducible to 0.06 per mil or better. If N2 is continually flowing into the source you will see 30 rise. But with time it should return down.
>>>
>>> It is my understanding that if heaters are off, water is still likely hanging around the source but you may not see a large m/z 18. Compared to leaving heaters on all the time, the water background will be much less.
>>>
>>> If your system doesn't seem stable, it may be worth it to close the inlet, turn on all heaters and let the system pump for a day or two then try again.
>>>
>>> Ben
>>>
>>> -----Original Message-----
>>> From: Stable Isotope Geochemistry <[log in to unmask]> On Behalf Of Arnold,Thomas E
>>> Sent: Tuesday, September 10, 2019 6:52 AM
>>> To: [log in to unmask]
>>> Subject: Re: [ISOGEOCHEM] Mass 30 instability on Delta V
>>>
>>> Ben (and others),
>>>
>>> Thanks for the quick response, we also assumed water in the system. As a follow-up: why would the m/z 30 signal increase throughout a single run and across multiple runs? You mention instability as a reason. What do you specifically mean by this? That the EA is not properly sealed? It’s also strange because we see this issue when the EA/high flow is disconnected from the IRMS and the on/off runs are through the conflo.
>>>
>>> Thanks again,
>>> TEA
>>>
>>>> On Sep 9, 2019, at 5:39 PM, Harlow, Benjamin <[log in to unmask]> wrote:
>>>>
>>>> Hi Tom,
>>>>
>>>> It is likely elevated water in your system. N from your N2 pulses combines with O from H2O to give you m/z 30. More time will help. We typically run with source and inlet valve heaters on all the time to keep water out. Sometimes a heat gun or hair dryer can be beneficial to drive water off other components. Monitor m/z 18 while heating various components. Watch for the signal to climb, eventually drop and stabilize.
>>>>
>>>> Every time you replace reagents or open up the EA more m/z 18 can get in your system. So venting the EA or allowing ample stability time is needed. If water is really bad then close the IRMS inlet and allow it to pump down independently.
>>>>
>>>> Usually if m/z 30 does the stair step on std on offs it is not a stable condition for us.
>>>>
>>>> Regards, Ben
>>>>
>>>> -----Original Message-----
>>>> From: Stable Isotope Geochemistry <[log in to unmask]> On Behalf Of Arnold,Thomas E
>>>> Sent: Monday, September 09, 2019 2:04 PM
>>>> To: [log in to unmask]
>>>> Subject: [ISOGEOCHEM] Mass 30 instability on Delta V
>>>>
>>>> To the great minds of the list serve,
>>>>
>>>> Our system: Delta V coupled to a costech EA via conflo IV
>>>>
>>>> Our issue: Mass 30 is drifting to higher and higher voltages across multiple N2 on/off runs — more than a 1V increase from our first to fourth on/off. I can send pictures to individual emails if needed.
>>>>
>>>> What we've tried thus far: 1)Replaced: the seals on the EA reactors, the copper wires in the reduction reactor, the water trap. 2) We turned on the heater to bake out the source and inlet. I believe this heater operates at a temperature of 50-80C. After the bake out, the amplitude of mass 30 was reduced, but the voltage drift was still there. 3) The problem presents itself in both EA mode and in the conflo only mode.
>>>>
>>>> Any suggestions on what to do? Should this even be considered a problem as our standard deviation looks fine after the 3rd series of on/off runs?
>>>> Thanks,
>>>> Elliott Arnold
>>>> Pitt Geology

Top of Message | Previous Page | Permalink

Advanced Options


Options

Log In

Log In

Get Password

Get Password


Search Archives

Search Archives


Subscribe or Unsubscribe

Subscribe or Unsubscribe


Archives

November 2020
October 2020
September 2020
August 2020
July 2020
June 2020
May 2020
April 2020
March 2020
February 2020
January 2020
December 2019
November 2019
October 2019
September 2019
August 2019
July 2019
June 2019
May 2019
April 2019
March 2019
February 2019
January 2019
December 2018
November 2018
October 2018
September 2018
August 2018
July 2018
June 2018
May 2018
April 2018
March 2018
February 2018
January 2018
December 2017
November 2017
October 2017
September 2017
August 2017
July 2017
June 2017
May 2017
April 2017
March 2017
February 2017
January 2017
December 2016
November 2016
October 2016
September 2016
August 2016
July 2016
June 2016
May 2016
April 2016
March 2016
February 2016
January 2016
December 2015
November 2015
October 2015
September 2015
August 2015
July 2015
June 2015
May 2015
April 2015
March 2015
February 2015
January 2015
December 2014
November 2014
October 2014
September 2014
August 2014
July 2014
June 2014
May 2014
April 2014
March 2014
February 2014
January 2014
December 2013
November 2013
October 2013
September 2013
August 2013
July 2013
June 2013
May 2013
April 2013
March 2013
February 2013
January 2013
December 2012
November 2012
October 2012
September 2012
August 2012
July 2012
June 2012
May 2012
April 2012
March 2012
February 2012
January 2012
December 2011
November 2011
October 2011
September 2011
August 2011
July 2011
June 2011
May 2011
April 2011
March 2011
February 2011
January 2011
December 2010
November 2010
October 2010
September 2010
August 2010
July 2010
June 2010
May 2010
April 2010
March 2010
February 2010
January 2010
December 2009
November 2009
October 2009
September 2009
August 2009
July 2009
June 2009
May 2009
April 2009
March 2009
February 2009
January 2009
December 2008
November 2008
October 2008
September 2008
August 2008
July 2008
June 2008
May 2008
April 2008
March 2008
February 2008
January 2008
December 2007
November 2007
October 2007
September 2007
August 2007
July 2007
June 2007
May 2007
April 2007
March 2007
February 2007
January 2007
December 2006
November 2006
October 2006
September 2006
August 2006
July 2006
June 2006
May 2006
April 2006
March 2006
February 2006
January 2006
December 2005
November 2005
October 2005
September 2005
August 2005
July 2005
June 2005
May 2005
April 2005
March 2005
February 2005
January 2005
December 2004
November 2004
October 2004
September 2004
August 2004
July 2004
June 2004
May 2004
April 2004
March 2004
February 2004
January 2004
December 2003
November 2003
October 2003
September 2003
August 2003
July 2003
June 2003
May 2003
April 2003
March 2003
February 2003
January 2003
December 2002
November 2002
October 2002
September 2002
August 2002
July 2002
June 2002
May 2002
April 2002
March 2002
February 2002
January 2002
December 2001
November 2001
October 2001
September 2001
August 2001
July 2001
June 2001
May 2001
April 2001
March 2001
February 2001
January 2001
December 2000
November 2000
October 2000
September 2000
August 2000
July 2000
June 2000
May 2000
April 2000
March 2000
February 2000
January 2000
December 1999
November 1999
October 1999
September 1999
August 1999
July 1999
June 1999
May 1999
April 1999
March 1999
February 1999
January 1999
December 1998
November 1998
October 1998
September 1998
August 1998
July 1998
June 1998
May 1998
April 1998
March 1998
February 1998
January 1998
December 1997
November 1997
October 1997
September 1997
August 1997
July 1997
June 1997
May 1997
April 1997
March 1997
February 1997
January 1997
December 1996
November 1996
October 1996
September 1996
August 1996
July 1996
June 1996
May 1996
April 1996
March 1996
February 1996
January 1996
December 1995
November 1995
October 1995
September 1995
August 1995
July 1995
June 1995
May 1995
April 1995

ATOM RSS1 RSS2



LIST.UVM.EDU

CataList Email List Search Powered by the LISTSERV Email List Manager