Hi All, We have a Delta Plus XL Mass Spectrometer that we vented last week to replace one of the quartz spacers in the source. While all the parts we laid on the table overnight (we had to order the spacer), the cooling unit in our building failed and the lab temp got up to 85 degrees. We had inadvertently left our cooling water turned after venting the instrument. When we reassembled the source and tried to pump down, the source vacuum will only pump down to 5X10-5. Thermo said that the source flange clamps were overtightened and the knife edge seal that seals the source housing needed to be replaced. Well we finally got the source flange clamps (from a secondary source), and replaced the knife edge seal and see that the source vacuum is still 10-5. The forevac vacuum is normal (2X10-3). My question(s): Could the temperature difference between the cooling water (15 degrees) and the air temperature (85 degrees) cause problems in the vented source housing and/or exposed flight tube? Could the high vacuum gauge be faulty? Do I suddenly not know how to tighten a source gasket to get the proper vacuum? Any ideas will be very much appreciated. Bernie Genna NCSU