Hi Kim, Assuming that the NA1500 is ready to go (Ready light is on), and that when you push the Start button the cycle timer begins to count up, the electrovalves for the autosampler should energize at the time set on the Sample Start thumbwheel switch. 1) The first thing to check is has anyone accidentally pushed down the "Sample" button in the Manual Exclusion section of the control panel (bottom row of buttons. That would inhibit the valves from being energized. 2) If the button is not depressed, check if the air is being switched from port 1 to port 2 at the time set for Sample Start. Remove the two blue pneumatic tubes which go to the autosampler, from ports 1 and 2. Air should only be coming out of line 1 if the EA is not running. Put your fingers over the two ports and start the NA 1500. If the air does not switch completely from port 1 to 2, and then back to line 1 (at the Sample Stop time), then either the valves are not being energized or there is a problem with the valves. 3) There is a simple test to see if the valves are being energized. Start the NA 1500, then after the Sample Start time touch a small screwdriver to the post which goes through the center of the solenoid. If the valve is energized, there will be a small magnetic attraction. 4) Next, check to see if the valves are leaking. There is a vent hole on the electrovalves which allows the air to vent from the autosampler air cylinder when pressure is applied to one side or the other. Locate the vent hole on the autosampler EVs (it's in the brass block), and see if any air is venting. There should be no air leak if the autosampler is not moving. Start the NA 1500 and when the valves are energized (Sample Start time) check to see if there is a continuous leak out of the vent when the air is switched from port 1 to port 2. If you find the valves are not being energized, there are a couple of possible causes. One of the wires to the solenoid may not be making good contact or the EVDV board is malfunctioning. There should be about 24 V across the solenoid when the valve is energized. If there is a continuous leak when air should be switched to line 2 (which moves the sampler forward), the most likely cause is a crack in the small plastic gas channeling part which is between the two valves. Instead of being channeled properly inside the valves, the air goes out the vent and you don't have enough pressure to drive the air piston forward. Best regards, Bruno Lavettre Costech Analytical Technologies, Inc. -----Original Message----- From: Stable Isotope Geochemistry [mailto:[log in to unmask]] On Behalf Of Kim Baublys Sent: Wednesday, January 18, 2006 5:44 PM To: [log in to unmask] Subject: more EA solenoid problems Hi All, We have a Carlo Erba NA1500 elemental analyser (attached to an Isoprime) and the auto sampler just stopped during an analytical run and will not operate even on manual. By switching the compressed air lines over I can get the auto-sampler to close, and then switch lines again and get it to open, which shows that the auto sampler is working OK. This led me to believe that the solenoid valve operating the auto sampler was faulty. I have tested all the coils and they have the same resistance (150ohms), so I am assuming that the solenoid valve is faulty. Either that or it is not getting the command to operate. Does this sound like a reasonable assumption? Has anyone else had this problem before? Is there anyway to test the solenoid valve without running the furnaces up to temperature? Any advice welcome. thanks kim