Hi Wendy, Peak at m/z=28 may result not only from N2, but also from CO background (provided that CO2 background is OK). Such increase from 5E-12 to 1.2E-11 seems to be normal after venting the ion source. Stan Halas At 15:58 06-03-13 -0500, you wrote: >Hi, Everyone, > > We are using Isoprime IRMS to run CO2 samples in DI mode. After the > ion source was cleaned, we run background scan (Nupro Valve Closed), the > N2 signal is ~12 e-12, it is much higher than before (0.8 e-12) and above > the acceptable value (5 e-12). However, the other signals, such as H2O, > O2, Ar, are not excess of their maximum acceptable values. > And to our experience, if there is a leak, Ar and H2O would show a > higher than normal background. Does anyone know what else may be causing > the high N2 background if it is not the leak? Any suggestion would be > greatly appreciated. > > >Wendy Zhang >Air Quality Research Division >Environment Canada >4905 Dufferin Street >Toronto, Ontario M3H 5T4 >Tel: (416) 739-5792 (office) >Tel: (416) 739-5822 (lab) >Fax: (416) 739-4664 >Email:[log in to unmask] > > Dr. Stanislaw Halas Mass Spectrometry Laboratory Institute of Physics, UMCS 20-031 Lublin, Poland phone: +48 81 537-62-75 fax: +48 81 533-36-69