A few weeks ago I had posted a message about our problem in hoping to get some tips from the community. The amount of help I got was overwhelming and I appreciate all that time you had spent on it.
The good news is the problem had been solved. It turned out, like some of you suggested, the power supply to the sampler was about to go. It works fine after the replacement.
The replacement was done by Mandell in their company (not the greatest experience, but purpose was served.) and costed us Can. $ 754.
Thanks to you all for the great help and I appreciate it.
----- Original Message -----
From: Li Huang <
[log in to unmask]>
Date: Thursday, May 24, 2007 3:48 pm
Subject: [ISOGEOCHEM] Gilson autosampler problem
To:
[log in to unmask]> Hi,
>
> We have had Multiprep for a few years and are experiencing some
> problem with Gilson autosampler. Here's what we see. In the
> middle of the autorun, every time the arm reaches row 3, 4 or 5
> ( our sampler has 5 row and 12 column.) Gilson returns to
> zero position right after the needle pulls out of the vial and
> will not move to the next position without manually initializing
> the Gilson.
>
> It is difficult for us to run autorun with the problem
> like this. It looks to me that there is problem on that section
> of the belt ( or track the sampler arm is travel on.) but don't
> know how to check it. I wonder if any body had seen this before
> and how they got this problem solved.
>
> Thanks,
>
> li
>
>
>
>
>
> Li Huang
>
> Laboratory for Stable Isotope Science
> University of Western Ontario
> London, Ontario
> Canada
> Phone: 1-519-661-3881
>
>