Marc, > I see vast nonlinearity when analyzing a water sample. Reference >and sample peaks look fine and sample peak intensities are as >expected for the sample weight. It seems like, in contrast to the >reference gas ratio calculations, the H3 factor is not included in >the ratio calculations for the samples. The technician and I have no >more ideas. It is the first time D/H is measured on this machine. > >First of all: Is the H3 correction for the samples commonly >calculated automatically by the mass spec software or do you have to >correct the sample ratios manually for the H3 contribution? >(Perhaps there is someone using the same isoprime mass spec and >could tell me about some hidden places in the Ion Vantage software >where a simple tick could solve all my problems...??) If the Ion Vantage software is similar to the MassLynx software used to run an Isoprime, there is indeed a box that needs a tick for the software to use the H3+ correction factor when calculating a final delta value. In MassLynx, you go to Edit Species, select H2, choose Reference Gas, and then the Molecule tab. In the H3+ Correction section, there's a box labelled "Enable" which needs to be checked for the calculations to use the H3+ correction factor. Incidentally, the "Use H3+ calibration data" box must have a check for the software to re-calculate the H3+ correction factor when you perform the routine to measure the H3+ correction factor. I haven't used or seen the Ion Vantage software, but I imagine that it has something similar. I hope this helps, and good luck! Cheers, Simon -- ***** Simon Poulson Research Associate Professor Dept. of Geological Sciences & Engineering MS-172 University of Nevada - Reno 1664 N. Virginia St. Reno, NV 89557-0138. USA (775) 784-1104 Fax: (775) 784-1833 [log in to unmask] http://www.mines.unr.edu/isotope/simonhome.html *****