1) You put in the del C13/C12 and del 18O/16O for your reference gas.
2) The software used standard del notation stuff to convert those to R(13C/12C) and R(18O/16O).
3) If you have configured it to use SSH it uses the equations from that paper (I have the reference around somewhere, write if you need it) to calculate R45 and R46 for your reference gas. I will refer to these as R45_knw & R46_ knw.
4) The raw numbers are used to construct the raw ratios rR45 (= rArea45/rArea44) and rR46 (=rArea46/rArea44). These raw numbers are already background corrected. I don't know too much about time shifts, they may be corrected for that as well. Note: this is also done for the rR45 and rR46 of the peak (or peaks) you have designated as reference peaks. Also note: by convention, lower case r stands for "raw" – this is an integer number that is in mVs, typically 5 integers long for my system. Do NOT use theVS numbers that contain decimals. I only use one reference peak for each run, so I'm not sure about how to use more than one. From here on out I assume 1. Anyway, for that one peak I will refer to rR45_mrf & rR46_mrf, where the mrf stands for measured reference )
5) In the spirit of internal standardization, calibration factors are created:
CF45 = R45_knw/rR45_mrf
CF46 = R46_knw/rR46_mrf
7) The SSH equations are then used to find R(13C/12C), R(18O/16O), R(17O/16O). (Note: if you use SSH, the equations are not analytical. I did them in MS Excel and used "Solver", you should probably do whatever you are most comfortable with – or just take it on faith.)
8) Standard delta equations are used …
This is what the reference peak does. It also is insensitive to minor variations in the multiplier feedback resistors, etc. Basically, this technique allows it to be a lot more precise. I think it is kind of cool, and I can't figure out why nobody wanted to tell me about it.
Patrick McLoughlin, Ph.D.
Technical Director
Microseeps, Inc.
==================================================
From Robert Schneider:
I thought about your question concerning the change of the R values when
changing the delta values of the reference gas. Also, I had the look to
the isodat software and reproduced your observation.
If you look on the 45R and 46R values, the same effect will take place,
but these values should be equal to one for the peak, that isodat uses
as reference peak (possibly peak one). For this peak, the 45R which is
45R = (Area_45/Area_44)/(Area_45_