Hi Mike- Sounds like you're sucking air... You may have to replumb it, but make sure you're getting He flow to the open splits--check both the He make up gas and the flow in from the GC. Occasionally we've had our capillaries catch and do a curly Q and suck air instead of going straight down. It's also possible you have have a crack in or inside of a ferrule. good luck... take care gerry Quoting "Kubo, Michael D. (ARC-SSX)[SETI INSTITUTE]" <[log in to unmask]>: > Hi Isogeochemers, > > After troubleshooting for a bit, I think my original diagnosis was > incorrect. It turns out that the pump HV source light kicks back on > when I closed the GCC valve, but didn't go out when I opened the > Conflo valve - just normal vacuum. I tried switching the capillaries > between the different valves and when they were switched, the results > were switched - improper vacuum on what was the Conflo valve (but now > had GCC capillaries going into it) and normal (2.2 x10^-7) on what > was the GCCIII valve (now with the conflo capillaries going into it). > > So there is clearly something wrong - probably a leak - with my > system upstream of the capillary/SGE valve interface. I tried > clipping about 3 cm of capillaries off of the inlet, in the hopes > that I had some kind of small crack right by the inlet nut on the > capillaries, but that didn't do anything. At this point, I'm not > quite sure how to approach this. My gut reaction would be to tear > out the capillary coming from the GCCIII interface and replumb it > into the source, but this seems a bit extreme. > > Any thoughts? > > Thanks again, > Mike Kubo > > > > ________________________________ > From: Stable Isotope Geochemistry [mailto:[log in to unmask]] On > Behalf Of Kubo, Michael D. (ARC-SSX)[SETI INSTITUTE] > Sent: Wednesday, July 29, 2009 1:23 PM > To: [log in to unmask] > Subject: [ISOGEOCHEM] Source HV Down on a ThermoFinnigan Delta Plus XL > > Hello Everyone, > > Of course, just before I'm about to run a 1 shot meteorite sample, > our HV just went down. Here is the play by play: > > 1) Had been running EA, so before I opened the GC-C-III SGE > valve, I turned off the filament HV button the front panel so as not > to flood the source with gas and trip the HV. > 2) Came back around to the front side of the instrument panel > to turn the source HV back on. Depressed button, only to see that > neither the source HV light or the emission light came on, but the > red LED on the button did. In addition, the source HV light on the > pump LED panel was also off. Source vacuum indicates 3.3 X 10^-5 - > clearly not the vacuum of 10^-6/10^-7 I typically operate at. > 3) If I try power cycling the "pump" switch on the front of the > instrument, the pump source HV LED will flash twice, then stay out. > > So my guess is that I have a bad source turbo and the source HV won't > kick on because of this. If anyone out there has other ideas that > don't involve weeks of downtime and several (tens of) thousands of > dollars to repair, PLEASE shoot me an email! Your help is greatly > appreciated (as always) and a lonely chunk of meteorite will be very > grateful as well. > > Cheers, > Mike Kubo > > > > **************************************************************************************************************************************** > Michael D. Kubo > NASA Ames Research Center > M/S 239-4 > Bldg. N239, Room 327 > Moffett Field, CA 94035 > > (650) 604-6110 > >