On the subject of cleaning syringes, I've found that phosphoric acid (any
acid, presumably) is great for rejuvenating 10uL syringes that seize up
with brown crud after many water injections.
Paul Eby
Alberta Research Council
At 10:33 AM 09/10/2009, you wrote:
Hi
Gael,
We have been using the LGR machines for a
while, and have switched away from the 1.2mcl syringe. The
plunger-in-needle style of syringe tends to lock after injecting water
and then sitting.
We moved to a 10mcl gas-tight syringe (SGE #002985) that has replaceable
parts (plunger and needle). At the time we switched, I do not think
that we were able to find a smaller gas-tight syringe with compareable
needle size. In order to get this syringe to work, you need to
change some of the settings. On the LGR machine, you need to change
the settings to the following:
1.1. Sample Volume = 150mcl
1.2. Standard Volume = 150mcl
1.3. Fill speed = 10,000nL/s
1.4. Injection speed = 10,000nL/s
1.5. Pre-injection delay = 1050ms
1.6. Post-injection delay =
50000ms
1.7. Post-withdrawal Delay =
10,000ms
The volumes and injection speeds are
assuming that you are still using the 1.2mcl syringe holder.
On the CTC autosampler, you need to change the pull-up delay to 1.2s and
the overfill to 2%. The only other issue is that if you use the
change syringe function on the autosampler, you crush the plunger. If you
use the 1.2mcl syringe, the autosampler will go to the correct plunger
height and then you can change to the 10mcl syringe without using the
change syringe function.
We have gotten good reproducibility with the number density using this
setup. This also allows us to run continually over the weekend
without the syringe getting "stuck." The cleaning
procedure would be to rinse 10 times with isopropanol, and then 10 times
with DI water. The first injection after cleaning tends to have a
high number density, but then settles down, which we get around by having
the first sample be DI water.
We tend to clean our transfer line/filter once a week by blowing
compressed air through the filter unit. Over time, you will get buildup
at the inlet side of the tubing, and we will either trim the transfer
line, or take off the filter and spray DI water thru the tubing to
dislodge the debris, followed by compressed air to get the water
out.
The septa we have run over the weekend without changing, but tend to
change it daily if time allows.
Hope that helps,
Tim
CONFIDENTIAL INFORMATION: The information contained in this email
and in any attachment is for use by the individual(s) to whom it is
addressed and shall be deemed “confidential information” within the
definition of any non-disclosure or confidentiality agreement between
KineMed, Inc. and the addressee or the company for which the addressee is
acting. Any unauthorized disclosure, copying, distribution, or use
of this information is prohibited.
From: Stable Isotope Geochemistry
[
mailto:[log in to unmask]] On Behalf Of Gael
MONVOISIN
Sent: Friday, October 09, 2009 4:47 AM
To: [log in to unmask]
Subject: [ISOGEOCHEM] Some LGR ICOS questions bis
Hi to you all isogeochem users.
Another list of questions about the LGR DLT 100 LWIA that we use for H/D
- 18O/16O measurements :
SIGNAL :
We ran our LGR without syringe, to check what the background was.
The signal decreases and I sent a picture of the screen to Rugh Williams
(thanks Rugh) who told me that the value on the y axis (transmitted
intensity) should be around 1.5 V. Our LGR response range is 0.5 V. Is it
only our LGR that “failed” or do other users work with such a transmitted
intensity in routine? Does it really can come only from a bad alignment
of the laser or dirty mirrors as Rugh suggested, or do someone have
another less frightening idea?
SYRINGUES :
Regarding the frequent problems with Hamilton 1.2 µl syringes and the
last email of Andrew Schauer about the cleaning procedure, I was
wondering to understand what would bring deposit or/and corrosion
of the piston, even if we inject only desionized or LGR standards
?
Andrew said he could get a very stable volume of injected water of
3.2*10^16. From our side, we are not stable at all, this value is moving
from 3.6 to 4.2*10^16. I guess this can't help to have good results if
the number of molecules is not stable? Would this lack of accuracy only
coming from the dirty syringes?
Does anyone can argue on the cleaning procedure frequency? How many runs
do you do with the same syringe before you change it?
Worse, recently we sent back 10 of a package of 12 brand new syringes to
Hamilton as the piston were not moving free in those new syringes. Did
other users ever had these problems ?
INJECTOR-TRANSFER LINE :
“Dust” of septa is produce by the syringe and introduce in the transfer
line. When do users clean the entire transfer line? Did you ever clean
the injector nut and how often? Change the cristal liner? Change the
filter? Or do you just blow some air in the line?
What is the effective number of injection that you consider before
any change of septa SGE ? On the LGR website they announced that
they doubled the number of injections, without changing the septa, on the
new LGR version ! The IAEA manual says to change it every 300 injections
more or less. After how many injections does the signal get worse and how
to be aware on that ?
Did anybody test others septa (I heard about thermagreen)?
CUSTOMISATION :
What about supplying the LGR with a N2 5.0 gas bottle instead of the
Drierite- dried lab atmosphere (for a drier and cleaner cell). Did
anybody ever tried that, what were the results and the input pressure you
selected ?
Does anybody changed the membran pump to a stronger one (decrease
of the memory effect ?). Is there a limit of pumping in the cell? Is it
possible to get troubles if the pump is too strong?
- About the septa of the vials, of the injector, did anybody tested
others septa (I heard about thermagreen but not tried it yet as we have
many troubles with our results)? What are the limits of the septa SGE of
LGR? On their website they doubled the number of injections on the new
LGR version, but without changing the septa! The IAEA manual says to
change it every 300 injections more or less. After how many injections
does the signal get worse and how can we see it?
USER COMMUNITY :
Is there a list of users of LWIA LGR? Shall we build one?
Gaël MONVOISIN
Ingénieur d'études CNRS
Service Hydrologie Isotopique
Laboratoire Interactions et Dynamiques des Environnements de Surface -
UMR 8148
Bâtiment 504 - 3° étage - Porte 326
Université Paris Sud - 91405 Orsay
France
00 33 (1) 69 15 71 74
email :
[log in to unmask]