We run single foram samples, normally 10-30 ug, on the Kiel IV + Thermo Delta V plus, with same size NBS19. Our long term precision is 0.05 for d13C and 0.10 for d18O, for small samples (size 10-30 ug). Even though I require sample size must be >10 ug to give reasonable delta values, sample size down to 8 ug also can yield enough gas for a reasonable analysis. Wei Huang, Ph.D. Stable Isotope Lab Manager New Core Lab Rm#212 Lamont-Doherty Earth Observatory of Columbia University 61 Route 9W Palisades, NY 10964 Lab: 845-365-8143 <845-365-8306> [log in to unmask] On Wed, Oct 25, 2017 at 12:24 PM, Alan Mix CEOAS <[log in to unmask]> wrote: > MAT252, Kiel-III, capillaries balanced at low voltage, cleanliness is next > to godliness, no slope corrections, “normal” glass reaction vials. > > Results to 10 micrograms on calcite standards — see graphic below (in the > 10-21 microgram range, d13C external precision of 0.03 to 0.04, d18O > external precision of 0.03 to 0.06 --two different numbers because two > different standards, offset between standards maintained at about 5.5 > permil for d18O, 1.2 permil for d13C, so no problem with gas mixing or > memory). Results on real biological samples will be less precise because > of, well, biology. > > I’m not particularly fond of approaches that that do slope corrections as > a function of mV as they can make standards look OK but you never really > know they are right. Not religious about this, but very cautious. Can’t > fight physics — if the capillaries jump flow regime at low pressures, > that’s the limit and the two sides don’t necessarily do it at exactly the > same point. On our current system, the practical limit appears to be about > 10 micrograms. > > Alan. > > > > ————— > This e-mail including any attachments is confidential and may be > priviledged. It is intended solely for the addressee. If you have received > it by mistake, please notify the sender by e-mail and delete this message > from your system. Any unauthorized use or dissemination of this e-mail in > whole or in part is strictly prohibited > ————— > Alan C. Mix, PhD > Distinguished Professor of Earth, Ocean, & Atmospheric Sciences > President of The Oceanography Society (www.tos.org) > Fellow of the American Association for the Advancement of Science > Fellow of the American Geophysical Union > > Oregon State University > CEOAS Administration Building 104 > Oregon State University > Corvallis, OR 97331-5503 USA > > email: [log in to unmask] <[log in to unmask]> > www: http://ceoas.oregonstate.edu/profile/mix/ > > > > On Oct 25, 2017, at 7:30 AM, Christopher Maupin <[log in to unmask]> > wrote: > > Dear Isogeofolks, a broad, community query here: > > I’m curious what your successes and/or failures running extremely small > carbonate samples have been like. Seeing the uptick in individual foram > papers in recent years as piqued my curiosity to reach out to the broader > iso-sphere. > > What instrument/peripherals have you tried/are you using? > What are the smallest masses of CaCO3 you are able to successfully measure? > What precision are you routinely achieving on these successful > measurements? > What is your “recipe” for success? > > Many thanks! > ________________________________________________ > Chris Maupin, Ph.D. > http://christophermaupin.strikingly.com > Research Associate and Facilities Manager > Stable Isotope Geosciences Facility (SIGF) > Geography Department > Texas A&M University > 405 Eller O&M Building 3147 TAMU > College Station, TX 77843 USA > Email: [log in to unmask], [log in to unmask] > Mobile: (512) 810-8437 > > > > > > >