Steve I agree with our collegues regarding the teflon valves. Some do tend to "breathe". Our carbonate vessels are a straight tube (Rotulex 19/9 Joint) with a side arm. The valves are ACE but with the glass stem. As for the vacuum greases we do use th Apiezon in the following way: Apiezon M: Day to day use at room temperature and for 25oC reaction of calcite. We use a very thin coat on the viton o-ring at the valve to reaction vessel seal and on the twin outer seals of the valve stem but no grease on the inner ring. Apiezon T: Is fine for 50oC for a period of 2 to 4 days. The M is too liquid at this temp. Apiezon H: Is good for longer reactions at 50oC such as magnesite (3 weeks min) with no atmospheric leakage. Apiezon N: The particularity of this one is that it seems to rubberize. It is good for long term seals such as vacuum traps at the pump. When attempting to remove it it "cracks" creating branching air leaks much like the silicon vac. grease.. However we don't use it anymore We tend now to only use the H and M. Almost all of our extraction techniques are greasless including the sample trapping (breakseals). This includes hydrogen reduction (Zn or Mn technique), CO2 and SO2 gases as well as MS connections. I hope this is of some help Gilles Gilles St-Jean Technologue de Recherche U d'Ottawa/ Géologie/ Laboratoire d'Isotope G.G. Hatch 140 Louis Pasteur, Ottawa, Ontario, Canada, K1N 6N5 Bureau/Office: 1-(613) 562-5800 ext. 6839 Lab: ext. 6836 Courier-E / E-Mail: [log in to unmask] -----Message d'origine----- De: Steve Nelson [SMTP:[log in to unmask]] Date: 2 juillet, 1997 14:19 À: [log in to unmask] Objet: high residual gas pressure Dear Listmembers, Once again I hope to enlist your help. I have built a conventional carbonate line and all seems to be working well now except one item--the reaction vessels--which seem to contain high residual gas pressures [about 40 microns upon expansion into the manifold]. The pressures are high enough that the gas transfer rates are slow. It takes 7-8 minutes to freeze down 100 micromoles of CO2 and clearly that is too long. This is after rough pumping overnight on the acid/sample and then pumping for at least a couple of hours with the diffusion pump and a cold trap before reaction. I have been letting samples react overnight. Once the CO2 is trapped and the residual gas is pumped away, the CO2 will freeze again very rapidly if I re-expand it. My two legged reaction vessel design was copied from a colleage. The upper half contains the teflon/o-ring stopcock and the lower half contains the two legs for reagent and sample. They are joined with a ACE o-ring connection and a clamp rather than a ground glass fitting. I ran blank vessels to determine whether the residual gas was related to the reagent or the reaction. I rough pumped on the vessels overnight and then with the diffusion pump and a cold trap, then I closed the stopcocks. I left the vessels on the manifold but let them sit overnight [assuming the problem is not in the stopcocks]. I also torched the bottom of the vessels before closing the stopcocks. Overnight a residual pressure of about 40 microns resulted. Hence, the source of the residual gas appears to be the reaction vessel itself. It seems to me that if the problem were leakage, that it would be unlikely that all of the vessels would produce such similar residual pressures or leak the same amount. Any ideas as to the cause or a possible solution would be appreciated. I've struggled with this for quite some time now. ******************************** Steve Nelson Dept. of Geology 673 WIDB Brigham Young University Provo. Utah 84602 voice: 1-801-378-8688 FAX: 1-801-378-8143 "INTEL INSIDE" is a warning label *********************************