Before checking for magnet orientation or loose or touching wires (and
they will move a tiny bit onto the springs when the source assembly is
inserted into the mass spec), a quick check is to just run a magnetic scan
across the range of masses the machine is designed for and see if you get
anything.



> I removed and cleaned the ion source for our delta V advantage. I changed
> a
> new filament and I put back the source and everything looks fine on isodat
> screen. Values areā€¦
>
> HV= 3.01KV
> Box= 0.72 mA
> Trap= 0.77 mA
> vacuum= 1.4e-006
> Magnet (steps)= 11572
>
> When I select reference gas (CO2) on the conflo screen to do a jump
> calibration no CO2 gas is flowing into the system. I checked for a leak
> and NO
> leak, we have a half tank CO2 gas. It seems ions are not reaching the cups
> or
> ions are not generated. I checked the CO2 tank is open, conflo is open and
> SGE valve is open.
>
> I am planning to remove the source again and see if the small opening in
> the
> side of the ion source chamber is somehow blocked or miss aligned (even
> though I checked every thing before I put it back) . Or when I pushed the
> source into the mass spectrometer possibly a wire may have been
> disconnected?.
>
> Any advice or suggestions before I remove the ion source again to check
> for
> any loose connection.
>
> Thanks
> Million
>