Hi Jason - I had a similar problem. My work around was to put a "dummy
peak" of ref gas CO after the spike from the peak jump but before the
unknown CO pulse, with a 3 second duration. The software would then
consistently recognize this dummy peak as distinct from the unknown CO
pulse consistently (mostly) so that the appropriate peaks could be set in
the method for the standards and unknown.

Maybe this is a bug in the software as when our TCEA was set up I was told
the software was not supposed to recognize the peak jump as a distinct
peak, yet it does.

Matt

On Monday, October 13, 2014, Jason Curtis <[log in to unmask]> wrote:

>  Hi All,
>
>  When running our new TC/EA with ConFlo IV into a Delta V Plus with
> Isodat 3 I get a very large spike right at the peak jump from H to CO.  The
> problem is that the software recognizes this as the start of the CO peak.
> The actual CO peak comes out something like 10-15 seconds after the peak
> jump.  As a result of this electronic noise I have to manually redefine the
> peak for every sample.
>
>  You can see a screen shot of what it looks like here:
> https://dl.dropboxusercontent.com/u/87418054/tcea spike.png
>
>  I can not see a method setting to stop the software from recognizing
> this spike as part of the sample peak.  I am sure that I am just missing
> something.  Any suggestions as to how to correct my method would be
> appreciated!
>
>  Thanks in advance,
> Jason
>
> Jason Curtis, Ph.D.
> Stable Isotope Mass Spec Lab manager
> Senior Associate-In Geochemistry
> Department of Geological Sciences
> 241 Williamson Hall
> University of Florida
> Gainesville, FL 32611
> [log in to unmask] <javascript:_e(%7B%7D,'cvml',[log in to unmask]);>, office
> 352-392-2296
> lab 352-392-3344, fax 352-392-9294
>
>
>
>
>