Thanks for the tips everyone. Turned out to be the GC setting for
handshaking.

Jeff



*__________________________________________________________________Jeff
Salacup, Ph.DLecturer **|** Stable Isotope & Biogeochemistry Lab Manager*
*UMass-Amherst*
Website: http://blogs.umass.edu/jsalacup/
 @https://www.researchgate.net/profile/Jeff_Salacup/
 @alkenones

*“… it’s not your life. It's life. Life is bigger than you… Life isn't
something that you possess; it's something that you take part in, and you
witness." –Louis C.K.*

On Tue, Oct 28, 2014 at 3:49 PM, Harlow, Benjamin <[log in to unmask]> wrote:

>  Hi Jeff,
>
>
>
> Check your handshaking.  I once had to reconfigure this when connecting a
> GC PAL to trace GC.  Here are the settings that were working for our trace
> and PAL:
>
>
>
> On keypad for the trace, hit configure and scroll down to handshaking.
>
>
>
> Remote start: High to Low
>
> Inhibit ready: when low
>
> end of run: high to low
>
>
>
> I hope this helps.
>
>
>
> Ben
>
>
>
>
>
> Benjamin Harlow
>
> Manager, Stable Isotope Core Laboratory
>
> Washington State University
>
> School of Biological Sciences
>
> 367 Eastlick
>
> Pullman, WA  99164-4236
>
>
>
> Office:  509-335-6161
>
> Lab:     509-335-6154
>
> Fax:     509-335-3184
>
>
>
> Laboratory for Biotechnology and Bioanalysis (LBB2) Stable Isotope Core
>
> http://www.isotopes.wsu.edu
>
>
>
>
>
> *From:* Stable Isotope Geochemistry [mailto:[log in to unmask]] *On
> Behalf Of *Mike Elefante, Elf Analytical Services
> *Sent:* Monday, October 27, 2014 4:56 PM
> *To:* [log in to unmask]
> *Subject:* Re: [ISOGEOCHEM] PAL Comm Issues
>
>
>
> Jeff,
>
>
>
>  If the cable from the pal to the GC is the correct one then you may only
> need to change the GC input for handshaking. The handshake port on the
> Trace uses pin 1 for remote start and pin 3 is ground. You can change the
> signal input of that port in the GC configuration. Hit config. key toggle
> down to handshake then you can change the signal input to accept either a
> high or low signal. It uses TTL input so range is 0 to 5 volts. I have
> never retrofitted the pal on the Trace but have set up many other devices
> and you can use the handshake or auto sampler port to accept a start signal
> from an external device. If you need more help I will be in my office
> tomorrow.
>
>
>
> Mike Elefante, President
>
> Elf Analytical Services, Inc
>
> 732-922-8400
>
>
>
>
>
>
>
> -----Original Message-----
> From: Jeff Salacup <[log in to unmask]>
> To: ISOGEOCHEM <[log in to unmask]>
> Sent: Mon, Oct 27, 2014 10:50 am
> Subject: [ISOGEOCHEM] PAL Comm Issues
>
> Hi collective Brain,
>
>
>
> This list has been so helpful over the past month or two that I'm coming
> back once again. In order to work around the Thermo AS3000 that was bending
> syringes, I've installed an older CTC Combi Pal, so much better. We are no
> longer bending syringes with blistering speed. However, the PAL will not
> tell the GC that the sample has been injected and that it should go ahead
> and start the oven temp program.
>
>
>
> So, Thermo Trace Ultra, comm cords between computer and GC and computer
> and PAL, as well as between GC and PAL (Interface 1 from PAL to Generic
> Handshake on GC). PAL starts injection program when I start a run from
> Acquisition in Isodat and knows where all the vials are. It just won't send
> the start signal to the GC. If I hit start everything goes as planned, with
> the PAL waiting until the GC starts cooling to run the injection program
> again, all on it's own.
>
>
>
> Any ideas?
>
>
>
> Thanks so much!!
>
>
>
>
>   Jeff
>
>
> *__________________________________________________________________ *
> *Jeff Salacup, Ph.D Lecturer | Stable Isotope & Biogeochemistry Lab
> Manager*
>
> *UMass-Amherst*
>
> Website: http://blogs.umass.edu/jsalacup/
>
>  @https://www.researchgate.net/profile/Jeff_Salacup/
>
>  @alkenones
>
>
>
> *“… it’s not your life. It's life. Life is bigger than you… Life isn't
> something that you possess; it's something that you take part in, and you
> witness." –Louis C.K.*
>