Print

Print


Thanks for the tips everyone. Turned out to be the GC setting for handshaking.

Jeff

__________________________________________________________________
Jeff Salacup, Ph.D
Lecturer 
| Stable Isotope & Biogeochemistry Lab Manager
UMass-Amherst
Website: http://blogs.umass.edu/jsalacup/
 @alkenones

“… it’s not your life. It's life. Life is bigger than you… Life isn't something that you possess; it's something that you take part in, and you witness." –Louis C.K.


On Tue, Oct 28, 2014 at 3:49 PM, Harlow, Benjamin <[log in to unmask]> wrote:

Hi Jeff,

 

Check your handshaking.  I once had to reconfigure this when connecting a GC PAL to trace GC.  Here are the settings that were working for our trace and PAL:

 

On keypad for the trace, hit configure and scroll down to handshaking.

 

Remote start: High to Low

Inhibit ready: when low

end of run: high to low

 

I hope this helps.

 

Ben

 

 

Benjamin Harlow

Manager, Stable Isotope Core Laboratory

Washington State University

School of Biological Sciences

367 Eastlick

Pullman, WA  99164-4236

 

Office:  509-335-6161

Lab:     509-335-6154

Fax:     509-335-3184

 

Laboratory for Biotechnology and Bioanalysis (LBB2) Stable Isotope Core

http://www.isotopes.wsu.edu

 

 

From: Stable Isotope Geochemistry [mailto:[log in to unmask]] On Behalf Of Mike Elefante, Elf Analytical Services
Sent: Monday, October 27, 2014 4:56 PM
To: [log in to unmask]
Subject: Re: [ISOGEOCHEM] PAL Comm Issues

 

Jeff,

 

 If the cable from the pal to the GC is the correct one then you may only need to change the GC input for handshaking. The handshake port on the Trace uses pin 1 for remote start and pin 3 is ground. You can change the signal input of that port in the GC configuration. Hit config. key toggle down to handshake then you can change the signal input to accept either a high or low signal. It uses TTL input so range is 0 to 5 volts. I have never retrofitted the pal on the Trace but have set up many other devices and you can use the handshake or auto sampler port to accept a start signal from an external device. If you need more help I will be in my office tomorrow.

 

Mike Elefante, President

Elf Analytical Services, Inc

 

 

 

-----Original Message-----
From: Jeff Salacup <[log in to unmask]>
To: ISOGEOCHEM <[log in to unmask]>
Sent: Mon, Oct 27, 2014 10:50 am
Subject: [ISOGEOCHEM] PAL Comm Issues

Hi collective Brain,

 

This list has been so helpful over the past month or two that I'm coming back once again. In order to work around the Thermo AS3000 that was bending syringes, I've installed an older CTC Combi Pal, so much better. We are no longer bending syringes with blistering speed. However, the PAL will not tell the GC that the sample has been injected and that it should go ahead and start the oven temp program.

 

So, Thermo Trace Ultra, comm cords between computer and GC and computer and PAL, as well as between GC and PAL (Interface 1 from PAL to Generic Handshake on GC). PAL starts injection program when I start a run from Acquisition in Isodat and knows where all the vials are. It just won't send the start signal to the GC. If I hit start everything goes as planned, with the PAL waiting until the GC starts cooling to run the injection program again, all on it's own. 

 

Any ideas?

 

Thanks so much!!

 


Jeff

__________________________________________________________________
Jeff Salacup, Ph.D
Lecturer | Stable Isotope & Biogeochemistry Lab Manager

UMass-Amherst

Website: http://blogs.umass.edu/jsalacup/

 @alkenones

 