Thanks to all that responded.

Thanks to the help of Bruno from Costech Analytical in the US we could identify a bad connection of the wires to the board that come from a thermistor that measures the temperature inside the power cabinet close the the transformer as the reason for the TRANS-T error (indicates that transformer temperature is >60C). The other suspects were the fans on the rear side of the insturment but these worked fine.

The results of this bad connection was a reading in the display of the instrument of +236C for the temperature of the transformer, which of course is above the limit of 60C and triggers the alarm. However, the reading stays identical without the thermistor connected to the board so we assumed either a burned/destroyed thermistor (infinity resistance) or simply a bad or no connection of the wires to the board. The resistance of the thermistor measured directly was fine (38kOhm) so we assume it was the latter. After putting everything back into place and ensuring good connection the readings are back to normal.

regards,
Robert


Am 28.11.2014 um 08:41 schrieb Robert van Geldern <[log in to unmask]>:

> Dear folks,
> 
> we run an Costech ECS 4010 elemental analyzer. It worked fine over the last 4 years but now we are getting more often an error message TRANS-T in the display during the run. The frequency of this error increased over the last months and now we receive in nearly every sequence. The manual has only very limited information about that but it seems that it gets a little warm inside (or at least the instrument thinks so). The display tells me that the reactors are hot but in reality they are cool and after a hard reset (switch EA off-on) the temperature readings are reasonable. So it seems that also the instrument software kind of freeze.
> 
> Sounds for me like a problem with the thermocouple or the power electronics but it seems we are the only Costech owners in Germany, so any help from the other side of the ocean is appreciated.
> 
> regards,
> Robert
> 
> -- 
> Dr. Robert van Geldern
> University of Erlangen-Nuremberg (FAU)
> Department of Geography and Geosciences
> GeoZentrum Nordbayern
> Schlossgarten 5
> 91054 Erlangen, Germany
> 
> [log in to unmask]
> www.gzn.uni-erlangen.de
> fon: +49-9131-85-22514
> fax: +49-9131-85-29294
> room: O 2.112 (office)

-- 
Dr. Robert van Geldern
University of Erlangen-Nuremberg (FAU)
Department of Geography and Geosciences
GeoZentrum Nordbayern
Schlossgarten 5
91054 Erlangen, Germany

[log in to unmask]
www.gzn.uni-erlangen.de
fon: +49-9131-85-22514
fax: +49-9131-85-29294
room: O 2.112 (office)