Jeff,

 

If your sample peaks show a size effect and your ref peaks don’t, then the answer could be in the reactor. If the conversion process is incomplete, you could easily get this result. You could try increasing the reactor temperature or looking at the whether the tube is sufficiently deactivated.

 

I’m also curious about the dilution series that you run: are you changing the concentration of the standard, or perhaps are you changing the split flow to change the peak height?

 

Paul Eby

 

 

Stable Isotope Lab Manager

InnoTech Alberta

 

TEL: 250-483-3290  CELL: 250-893-5920
[log in to unmask]

 

Vancouver Island Technology Park
#3-4476 Markham St., Victoria
BC, Canada  V8Z 7X8

 

InnoTechAlberta.ca

 

This communication may contain confidential, personal and/or privileged information. Please immediately contact the sender if you are not the intended recipient of this communication. If you are not the intended recipient of this communication, disclosure, copying, distribution, use and/or reliance of this communication is strictly prohibited. Any communication received in error, or subsequent reply, should immediately be deleted or destroyed.

 

 

 

From: Stable Isotope Geochemistry [mailto:[log in to unmask]] On Behalf Of Jeff Salacup
Sent: January-17-17 8:07 AM
To: [log in to unmask]
Subject: Re: [ISOGEOCHEM] dD and linearity after H3

 

Thanks, Matheus.

 

I've got an older GCC III that I have to manually adjust.

 

So, if I do a series of on/offs, 8 refgas peaks, of increasing size from 1 V to 10 V I get a much smaller effect of ~ 0.3 permil/V with an H3 of ~ 3.2 applied. This same H3 is applied to my alkane standards, but I still have the effect I described above. 

 

Grumble...