Hi Josie,
Standards for our lab....
Time:....
1
hour for ABI / ABI and EX (the other techs get 1 hour and 30 min...can
u believe it? but I tell them to book them 1 hour in my room)
1
hour for Unilat LEA
1
hour 30 min for Bilat LEA
(our
schedulers do not understand our testing and so it cannot get more specific
than that)
A
basic ABI /PPG takes 30 min with NO ex but our new tech takes 1 hour.
A
graft imaging takes 20 min but these are rarely booked correctly.
I
can do an ABI and Unilat graft imaging in one hour, maybe less but it is
hard to depend on the schedulers...arterial imaging takes longer than graft
imaging for sure.
Scanning Protocol...
Our docs tried to give me a hard time
about doing the left leg before the right but honestly, I am getting old
and my arm is tired after one leg so I want the leg I finish with to be
closest to me (much easier to scan), same with Iliacs and Venous Studies.
The doc couldn't really argue with that. We hurt ourselves
each and everyday so this is one way I adapt. I have learned to scan
with my left arm so I do all unilat left side studies with my left arm
but bilateral studies get done with my right. I scan much slower
with my left and would run out of time if I scanned left. I am not
saying it is the right thing to do but it is my explanation. Other
than that I cannot see why someone would do the left first. It keeps
the docs on their toes reading reports lol.
Measurements...
This has recently been brought to my
attention. The employees change so often that I often feel I am the
only one who remembers when the protocols were enacted lol. I keep
doing things the same and as new employees come in they do things their
way. After awhile people are all doing things differently. I
measure and record TRV AP x TRV Width x Long AP for Aneurysm studies, this
was agreed on years ago. They may soon change how it is recorded
because we have a new Tech Director but we will see.
Not sure if this helps...Take care,
Sincerely,
Nicole Ball, BS, RVT
President for NNEVS
Frisbie Memorial Hospital
Vascular Lab
Rochester, NH
(w) 603-332-3100 ext 8127
From:
Lisa Kincaid <[log in to unmask]>
To:
[log in to unmask]
Date:
12/28/2016 09:54 AM
Subject:
Standards
Sent by:
UVM Flownet
<[log in to unmask]>
Just a note :
We still allow 2 hrs for a LE arterial. I can do in an hour , but
I have a lot of experience. I will squeeze in an arterial and even do within
a 1/2hour, but not general rule.
There are still those patients that require multiple "potty"
breaks, assistance with dressing and in and out of wheelchairs, and just
general caring needs. We have a lot of woundcare patients, and they are
time consuming in general.
This is why we allow extra time. It is not all about the sonographer. We
have great reporting software but still takes time to download the information
into the report.
We are a small lab in a surgeon's office. This is a lot different than
the 20 years I put in at the hospital and research. But I have noticed
the hospital studies are abbreviated since I left (and there are reasons
for this also).
The "extra time" that may be available to a sonographer is used
to get studies ready for the next day, catch up on studies they couldn't
write up before the next patient, and add ons ,ect.
I am very lucky to work with a surgeons who cares as much about their sonographers
as their patients. We try very hard to avoid work injuries and repetitive
injuries, thereby keeping our consistency and standards high. And
as a benefit, patient satisfaction.
Lisa Kincaid, RVT
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of UVMFLOWNET automatic digest system
Sent: Tuesday, December 27, 2016 21:00 PM
To: [log in to unmask]
Subject: UVMFLOWNET Digest - 24 Dec 2016 to 27 Dec 2016 (#2016-190)
There are 6 messages totaling 2477 lines in this issue.
Topics of the day:
1. DEVELOPED STANDARDS FOR IMAGING (6)
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
----------------------------------------------------------------------
Date: Tue, 27 Dec 2016 17:51:30 +0000
From: "Douville, Colleen" <[log in to unmask]>
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Andrew<
Which reporting software are you using?
Colleen Douville
Director l Vascular Ultrasound & Neurophysiology
[log in to unmask]
Office:
206-320-4080
Cell:
206-601-8077
Swedish Neuroscience Institute
500 17th Ave l 4 West, Room 400
Seattle l WA l 98122
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Bachman, Andrew
Sent: Friday, December 23, 2016 8:02 AM
To: [log in to unmask]
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Hi Josie,
I can not answer what the 'standard' is across the board because there
are some labs that go above and beyond, some labs that do the bare
minimum and hundreds in between. I work for the largest heart and
vascular center in the country currently (by sheer number of locations
and studies, and techs) and I will answer based on our practices but we
by no means set the standards for our field.
1- Measuring:
- I have always seen vascular measuring AP
x TRV but this could be a regional issue or a difference in whom is reading
studies (Surgeons vs Radiologists). We have had issues with dual
general/vascular technologists giving push back on this in years past.
- If everything is consistent and uniform across
the board with all techs, this should not be an issue but 1 person differs
and you will create a ton of confusion to both ordering doctors and reading
physicians.
2- Imaging order/standard
- This depends slightly on your reporting software.
- We use a software that allows our doctors
to click on a segment of the artery and pull up only the images for that
location allowing them to see diseased areas side by side with previous
images. In our case, order is never a problem and techs are free
to scan right then left, or left then right, or any variation.
- If your physicians review each individual
image, then you will need a standard across the board for everyone but
I think in the grand scheme of things, WHAT IMAGES YOU TAKE is more important
then the order you take them
- We do on average 150-200 studies a day with
5 surgeons reading daily, and 5 radiologists reading a few times a month.
If we didn’t have our reporting system, we would 100% have a standard
image order or we would never function.
3- Timeframe
- This again depends on your u/s equipment,
ABI machine and software. Years ago the standard was 1.5hours for
an LEA but the older machines required pressing 5 buttons to change 1 thing
and 15 different tweaks to make a pretty image.
- We have all new
GE machines, newer parks machines (at majority of our locations), and techs
do a short H&P with a full detailed report at completion. We
perform ABI's and B/L Arterial with the completed report in 1 hour but
in some of our locations we have 1.5hrs for B/L due to dual techs and older
machines. Unilateral studies are always 1hour or less regardless
of the location/machine.
- I have worked
in labs with older machines and you do need that extra 30mins at times
for bilateral studies or difficult/challenging patients.
- For a Unilateral study (especially simplistic bypass graft
such as a fem-pop), they should not need more then 1 hour.
- I would recommend monitoring their studies. See if
they are using that entire 1.5 hours or if they are finishing early and
lounging for 30mins.
- All this aside, I have also, in the past,
worked in a VERY respectable lab with old HDI-5000 machines and we did
ABI's, B/L Arterial and AortoIliac all in 1 hour. Plus we did the
reporting. So I guess even by old practice standards I am reluctant
to really say 1.5hrs is 'needed'.
I hope this helps.
Andrew M. Bachman, Bs RVT
St. Lukes University Hospital and Health Network
The Heart and Vascular Center
"To handle yourself, use your head. To handle others, use your heart"
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Josie Klapec
Sent: Friday, December 23, 2016 10:15 AM
To: [log in to unmask]
Subject: DEVELOPED STANDARDS FOR IMAGING
Hello Everyone -
I posted this on the SVU site and got one response. I understand
that the questions are a bit basic but everyone's opinions are welcome.
I am updating protocols at a new job. They haven't been updated since
2001.
The technologists were taught to always measure vessels TRV x AP.
New students and myself (not new by any stretch of the imagination!)
have been taught to measure vessels (not masses) as AP x TRV. I
need to standardize and am wondering if this is an industry standard?
Also, the team has not had an effective supervisor for some time (4-5
years) Anarchy is rampant. Some scan the left first and others
scan the right. Once again, new students and myself have been taught
to always start on the right, (disregard extenuating circumstances). Any
standardization is met with a lot of criticism. I have repeated reproducibility,
surgeons wishes, etc.
Finally, I have gotten a lot of push back from them stating they are unable
to complete a unilateral fem-pop bypass graft and ABIs in one hour. How
much time do you allow?
I have reference both IAC and SVU standards/guidelines. This is not
a small lab - I have 22 direct reports. Any thoughts are greatly
appreciated!! Quite a bit to respond to, but please, if it is slow,
answer away!
Josie Klapec, BS, RVT
To unsubscribe or search other topics on UVM Flownet link to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__list.uvm.edu_archives_uvmflownet.html&d=DQIBaQ&c=kLuyfJs8xHieIsti43i5Bw&r=nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=EX5MA9x3D7W_q8SmeNCaBmdGYHFx3OYgY4gB2s5y62E&s=5XGuKWYeJIQGlMptsioOsUa71N5E8TiRWxZcpvTCMyI&e=
[St. Luke’s University Hospital is a 4-time recipient of this prestigious
award – Click the image to learn more]<http://www.slhn.org/top100>
Confidentiality Notice: This e-mail message, including any attachments,
is for the sole use of intended recipient(s) and may contain confidential
and privileged information. Any unauthorized review, use, disclosure or
distribution is prohibited. If you are not the intended recipient, please
contact the sender by reply e-mail and destroy all copies of the original
message.
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
------------------------------
Date: Tue, 27 Dec 2016 18:01:26 +0000
From: "Bachman, Andrew" <[log in to unmask]>
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Colleen,
Please email me on my work email and we can discuss this. We love
our software but don’t want to start a software battle in this email chain.
[log in to unmask]
Andrew M. Bachman, Bs RVT
St. Lukes University Hospital and Health Network
The Heart and Vascular Center
"To handle yourself, use your head. To handle others, use your heart"
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Douville, Colleen
Sent: Tuesday, December 27, 2016 12:52 PM
To: [log in to unmask]
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Andrew<
Which reporting software are you using?
Colleen Douville
Director l Vascular Ultrasound & Neurophysiology [log in to unmask]
Office:
206-320-4080
Cell:
206-601-8077
Swedish Neuroscience Institute
500 17th Ave l 4 West, Room 400
Seattle l WA l 98122
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Bachman, Andrew
Sent: Friday, December 23, 2016 8:02 AM
To: [log in to unmask]
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Hi Josie,
I can not answer what the 'standard' is across the board because there
are some labs that go above and beyond, some labs that do the bare
minimum and hundreds in between. I work for the largest heart and
vascular center in the country currently (by sheer number of locations
and studies, and techs) and I will answer based on our practices but we
by no means set the standards for our field.
1- Measuring:
- I have always seen vascular measuring AP
x TRV but this could be a regional issue or a difference in whom is reading
studies (Surgeons vs Radiologists). We have had issues with dual
general/vascular technologists giving push back on this in years past.
- If everything is consistent and uniform across
the board with all techs, this should not be an issue but 1 person differs
and you will create a ton of confusion to both ordering doctors and reading
physicians.
2- Imaging order/standard
- This depends slightly on your reporting software.
- We use a software that allows our doctors
to click on a segment of the artery and pull up only the images for that
location allowing them to see diseased areas side by side with previous
images. In our case, order is never a problem and techs are free
to scan right then left, or left then right, or any variation.
- If your physicians review each individual
image, then you will need a standard across the board for everyone but
I think in the grand scheme of things, WHAT IMAGES YOU TAKE is more important
then the order you take them
- We do on average 150-200 studies a day with
5 surgeons reading daily, and 5 radiologists reading a few times a month.
If we didn’t have our reporting system, we would 100% have a standard
image order or we would never function.
3- Timeframe
- This again depends on your u/s equipment,
ABI machine and software. Years ago the standard was 1.5hours for
an LEA but the older machines required pressing 5 buttons to change 1 thing
and 15 different tweaks to make a pretty image.
- We have all new
GE machines, newer parks machines (at majority of our locations), and techs
do a short H&P with a full detailed report at completion. We
perform ABI's and B/L Arterial with the completed report in 1 hour but
in some of our locations we have 1.5hrs for B/L due to dual techs and older
machines. Unilateral studies are always 1hour or less regardless
of the location/machine.
- I have worked
in labs with older machines and you do need that extra 30mins at times
for bilateral studies or difficult/challenging patients.
- For a Unilateral study (especially simplistic bypass graft
such as a fem-pop), they should not need more then 1 hour.
- I would recommend monitoring their studies. See if
they are using that entire 1.5 hours or if they are finishing early and
lounging for 30mins.
- All this aside, I have also, in the past,
worked in a VERY respectable lab with old HDI-5000 machines and we did
ABI's, B/L Arterial and AortoIliac all in 1 hour. Plus we did the
reporting. So I guess even by old practice standards I am reluctant
to really say 1.5hrs is 'needed'.
I hope this helps.
Andrew M. Bachman, Bs RVT
St. Lukes University Hospital and Health Network The Heart and Vascular
Center
"To handle yourself, use your head. To handle others, use your heart"
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Josie Klapec
Sent: Friday, December 23, 2016 10:15 AM
To: [log in to unmask]
Subject: DEVELOPED STANDARDS FOR IMAGING
Hello Everyone -
I posted this on the SVU site and got one response. I understand
that the questions are a bit basic but everyone's opinions are welcome.
I am updating protocols at a new job. They haven't been updated since
2001.
The technologists were taught to always measure vessels TRV x AP.
New students and myself (not new by any stretch of the imagination!)
have been taught to measure vessels (not masses) as AP x TRV. I
need to standardize and am wondering if this is an industry standard?
Also, the team has not had an effective supervisor for some time (4-5
years) Anarchy is rampant. Some scan the left first and others
scan the right. Once again, new students and myself have been taught
to always start on the right, (disregard extenuating circumstances). Any
standardization is met with a lot of criticism. I have repeated reproducibility,
surgeons wishes, etc.
Finally, I have gotten a lot of push back from them stating they are unable
to complete a unilateral fem-pop bypass graft and ABIs in one hour. How
much time do you allow?
I have reference both IAC and SVU standards/guidelines. This is not
a small lab - I have 22 direct reports. Any thoughts are greatly
appreciated!! Quite a bit to respond to, but please, if it is slow,
answer away!
Josie Klapec, BS, RVT
To unsubscribe or search other topics on UVM Flownet link to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__list.uvm.edu_archives_uvmflownet.html&d=DQIBaQ&c=kLuyfJs8xHieIsti43i5Bw&r=nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=EX5MA9x3D7W_q8SmeNCaBmdGYHFx3OYgY4gB2s5y62E&s=5XGuKWYeJIQGlMptsioOsUa71N5E8TiRWxZcpvTCMyI&e=
[St. Luke’s University Hospital is a 4-time recipient of this prestigious
award – Click the image to learn more]<http://www.slhn.org/top100>
Confidentiality Notice: This e-mail message, including any attachments,
is for the sole use of intended recipient(s) and may contain confidential
and privileged information. Any unauthorized review, use, disclosure or
distribution is prohibited. If you are not the intended recipient, please
contact the sender by reply e-mail and destroy all copies of the original
message.
To unsubscribe or search other topics on UVM Flownet link to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__list.uvm.edu_archives_uvmflownet.html&d=DQIDaQ&c=kLuyfJs8xHieIsti43i5Bw&r=nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=vD_EIjFKGl9cMgs4mwlhhWwzj8ldS5jKwRLNhLwEIY8&s=T-IUflS8MOd58Ea_lOwz8qZc0y0-UlgN7LWTDUM2QbY&e=
To unsubscribe or search other topics on UVM Flownet link to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__list.uvm.edu_archives_uvmflownet.html&d=DQIDaQ&c=kLuyfJs8xHieIsti43i5Bw&r=nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=vD_EIjFKGl9cMgs4mwlhhWwzj8ldS5jKwRLNhLwEIY8&s=T-IUflS8MOd58Ea_lOwz8qZc0y0-UlgN7LWTDUM2QbY&e=
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
------------------------------
Date: Tue, 27 Dec 2016 13:49:15 -0800
From: Steve Knight <[log in to unmask]>
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Josie, when you say AP X TRV, do you mean AP in a longitudinal plane or
AP
in a transverse plane? AP in a transverse plane is prone to error due to
misalignment (particularly in the proximal aorta). TRV in the transverse
plane is also prone to error due to poor resolution of the lateral walls
and misalignment.
I like to use AP in a long plane. I will also measure in TRV for comparison
and they should be close but I trust the AP long more.
I know my stance on this is not the popular one but I believe my reasons
are valid.
~S
On Tue, Dec 27, 2016 at 10:01 AM, Bachman, Andrew <[log in to unmask]>
wrote:
> Colleen,
> Please email me on my work email and we can discuss this. We
love our
> software but don’t want to start a software battle in this email
chain.
>
> [log in to unmask]
>
> Andrew M. Bachman, Bs RVT
> St. Lukes University Hospital and Health Network
> The Heart and Vascular Center
>
> "To handle yourself, use your head. To handle others, use your
heart"
>
>
>
> -----Original Message-----
> From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Douville,
> Colleen
> Sent: Tuesday, December 27, 2016 12:52 PM
> To: [log in to unmask]
> Subject: Re: DEVELOPED STANDARDS FOR IMAGING
>
> Andrew<
> Which reporting software are you using?
>
>
> Colleen Douville
> Director l Vascular Ultrasound & Neurophysiology
> [log in to unmask]
> Office: 206-320-4080
> Cell: 206-601-8077
>
> Swedish Neuroscience Institute
> 500 17th Ave l 4 West, Room 400
> Seattle l WA l 98122
>
>
>
>
> -----Original Message-----
> From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Bachman,
> Andrew
> Sent: Friday, December 23, 2016 8:02 AM
> To: [log in to unmask]
> Subject: Re: DEVELOPED STANDARDS FOR IMAGING
>
> Hi Josie,
> I can not answer what the 'standard' is across the board because there
are
> some labs that go above and beyond, some labs that do the bare
minimum and
> hundreds in between. I work for the largest heart and vascular
center in
> the country currently (by sheer number of locations and studies, and
techs)
> and I will answer based on our practices but we by no means set the
> standards for our field.
>
> 1- Measuring:
> - I have always seen vascular measuring
AP x TRV but this could be
> a regional issue or a difference in whom is reading studies (Surgeons
vs
> Radiologists). We have had issues with dual general/vascular
technologists
> giving push back on this in years past.
> - If everything is consistent and uniform
across the board with
> all techs, this should not be an issue but 1 person differs and you
will
> create a ton of confusion to both ordering doctors and reading physicians.
> 2- Imaging order/standard
> - This depends slightly on your reporting
software.
> - We use a software that allows our doctors
to click on a segment
> of the artery and pull up only the images for that location allowing
them
> to see diseased areas side by side with previous images. In
our case,
> order is never a problem and techs are free to scan right then left,
or
> left then right, or any variation.
> - If your physicians review each individual
image, then you will
> need a standard across the board for everyone but I think in the grand
> scheme of things, WHAT IMAGES YOU TAKE is more important then the
order you
> take them
> - We do on average 150-200 studies a day
with 5 surgeons reading
> daily, and 5 radiologists reading a few times a month. If we
didn’t have
> our reporting system, we would 100% have a standard image order or
we would
> never function.
> 3- Timeframe
> - This again depends on your u/s equipment,
ABI machine and
> software. Years ago the standard was 1.5hours for an LEA but
the older
> machines required pressing 5 buttons to change 1 thing and 15 different
> tweaks to make a pretty image.
> - We have
all new GE machines, newer parks machines (at
> majority of our locations), and techs do a short H&P with a full
detailed
> report at completion. We perform ABI's and B/L Arterial with
the completed
> report in 1 hour but in some of our locations we have 1.5hrs for B/L
due to
> dual techs and older machines. Unilateral studies are always
1hour or less
> regardless of the location/machine.
> - I have worked
in labs with older machines and you do
> need that extra 30mins at times for bilateral studies or
> difficult/challenging patients.
>
- For a Unilateral study (especially simplistic
> bypass graft such as a fem-pop), they should not need more then 1
hour.
>
- I would recommend monitoring their studies. See
> if they are using that entire 1.5 hours or if they are finishing early
and
> lounging for 30mins.
> - All this aside, I have also, in the
past, worked in a VERY
> respectable lab with old HDI-5000 machines and we did ABI's, B/L Arterial
> and AortoIliac all in 1 hour. Plus we did the reporting. So
I guess even
> by old practice standards I am reluctant to really say 1.5hrs is 'needed'.
>
> I hope this helps.
> Andrew M. Bachman, Bs RVT
> St. Lukes University Hospital and Health Network The Heart and Vascular
> Center
>
> "To handle yourself, use your head. To handle others, use your
heart"
>
>
>
>
> -----Original Message-----
> From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Josie
> Klapec
> Sent: Friday, December 23, 2016 10:15 AM
> To: [log in to unmask]
> Subject: DEVELOPED STANDARDS FOR IMAGING
>
> Hello Everyone -
>
> I posted this on the SVU site and got one response. I understand
that the
> questions are a bit basic but everyone's opinions are welcome.
>
> I am updating protocols at a new job. They haven't been updated
since
> 2001.
>
> The technologists were taught to always measure vessels TRV x AP.
> New students and myself (not new by any stretch of the imagination!)
> have been taught to measure vessels (not masses) as AP x TRV.
I
> need to standardize and am wondering if this is an industry standard?
>
> Also, the team has not had an effective supervisor for some time (4-5
> years) Anarchy is rampant. Some scan the left first and
others scan the
> right. Once again, new students and myself have been taught
to always
> start on the right, (disregard extenuating circumstances). Any
> standardization is met with a lot of criticism. I have repeated
> reproducibility, surgeons wishes, etc.
>
> Finally, I have gotten a lot of push back from them stating they are
> unable to complete a unilateral fem-pop bypass graft and ABIs in one
hour.
> How much time do you allow?
>
> I have reference both IAC and SVU standards/guidelines. This
is not a
> small lab - I have 22 direct reports. Any thoughts are greatly
> appreciated!! Quite a bit to respond to, but please, if it is
slow, answer
> away!
>
> Josie Klapec, BS, RVT
>
> To unsubscribe or search other topics on UVM Flownet link to:
> https://urldefense.proofpoint.com/v2/url?u=http-3A__list.
> uvm.edu_archives_uvmflownet.html&d=DQIBaQ&c=kLuyfJs8xHieIsti43i5Bw&r=
> nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=EX5MA9x3D7W_
> q8SmeNCaBmdGYHFx3OYgY4gB2s5y62E&s=5XGuKWYeJIQGlMptsioOsUa71N5E8T
> iRWxZcpvTCMyI&e=
>
> [St. Luke’s University Hospital is a 4-time recipient of this prestigious
> award – Click the image to learn more]<http://www.slhn.org/top100>
>
> Confidentiality Notice: This e-mail message, including any attachments,
is
> for the sole use of intended recipient(s) and may contain confidential
and
> privileged information. Any unauthorized review, use, disclosure or
> distribution is prohibited. If you are not the intended recipient,
please
> contact the sender by reply e-mail and destroy all copies of the original
> message.
>
> To unsubscribe or search other topics on UVM Flownet link to:
> https://urldefense.proofpoint.com/v2/url?u=http-3A__list.
> uvm.edu_archives_uvmflownet.html&d=DQIDaQ&c=kLuyfJs8xHieIsti43i5Bw&r=
> nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=vD_
> EIjFKGl9cMgs4mwlhhWwzj8ldS5jKwRLNhLwEIY8&s=T-IUflS8MOd58Ea_
> lOwz8qZc0y0-UlgN7LWTDUM2QbY&e=
>
> To unsubscribe or search other topics on UVM Flownet link to:
> https://urldefense.proofpoint.com/v2/url?u=http-3A__list.
> uvm.edu_archives_uvmflownet.html&d=DQIDaQ&c=kLuyfJs8xHieIsti43i5Bw&r=
> nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=vD_
> EIjFKGl9cMgs4mwlhhWwzj8ldS5jKwRLNhLwEIY8&s=T-IUflS8MOd58Ea_
> lOwz8qZc0y0-UlgN7LWTDUM2QbY&e=
>
> To unsubscribe or search other topics on UVM Flownet link to:
> http://list.uvm.edu/archives/uvmflownet.html
>
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
------------------------------
Date: Tue, 27 Dec 2016 23:50:57 +0000
From: Derek Butler <[log in to unmask]>
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Steve –
Had to comment – I love this statement: “I know my stance on this is
not the popular one but I believe my reasons are valid.” – I wonder if
that will work at home, lolol.
Derek -
From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Steve Knight
Sent: Tuesday, December 27, 2016 4:49 PM
To: [log in to unmask]
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Josie, when you say AP X TRV, do you mean AP in a longitudinal plane or
AP in a transverse plane? AP in a transverse plane is prone to error due
to misalignment (particularly in the proximal aorta). TRV in the transverse
plane is also prone to error due to poor resolution of the lateral walls
and misalignment.
I like to use AP in a long plane. I will also measure in TRV for comparison
and they should be close but I trust the AP long more.
I know my stance on this is not the popular one but I believe my reasons
are valid.
~S
On Tue, Dec 27, 2016 at 10:01 AM, Bachman, Andrew <[log in to unmask]<mailto:[log in to unmask]>>
wrote:
Colleen,
Please email me on my work email and we can discuss this. We love
our software but don’t want to start a software battle in this email chain.
[log in to unmask]<mailto:[log in to unmask]>
Andrew M. Bachman, Bs RVT
St. Lukes University Hospital and Health Network
The Heart and Vascular Center
"To handle yourself, use your head. To handle others, use your heart"
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]<mailto:[log in to unmask]>]
On Behalf Of Douville, Colleen
Sent: Tuesday, December 27, 2016 12:52 PM
To: [log in to unmask]<mailto:[log in to unmask]>
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Andrew<
Which reporting software are you using?
Colleen Douville
Director l Vascular Ultrasound & Neurophysiology [log in to unmask]<mailto:[log in to unmask]>
Office: 206-320-4080<tel:206-320-4080>
Cell: 206-601-8077<tel:206-601-8077>
Swedish Neuroscience Institute
500 17th Ave l 4 West, Room 400
Seattle l WA l 98122
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]<mailto:[log in to unmask]>]
On Behalf Of Bachman, Andrew
Sent: Friday, December 23, 2016 8:02 AM
To: [log in to unmask]<mailto:[log in to unmask]>
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Hi Josie,
I can not answer what the 'standard' is across the board because there
are some labs that go above and beyond, some labs that do the bare
minimum and hundreds in between. I work for the largest heart and
vascular center in the country currently (by sheer number of locations
and studies, and techs) and I will answer based on our practices but we
by no means set the standards for our field.
1- Measuring:
- I have always seen vascular measuring AP
x TRV but this could be a regional issue or a difference in whom is reading
studies (Surgeons vs Radiologists). We have had issues with dual
general/vascular technologists giving push back on this in years past.
- If everything is consistent and uniform across
the board with all techs, this should not be an issue but 1 person differs
and you will create a ton of confusion to both ordering doctors and reading
physicians.
2- Imaging order/standard
- This depends slightly on your reporting software.
- We use a software that allows our doctors
to click on a segment of the artery and pull up only the images for that
location allowing them to see diseased areas side by side with previous
images. In our case, order is never a problem and techs are free
to scan right then left, or left then right, or any variation.
- If your physicians review each individual
image, then you will need a standard across the board for everyone but
I think in the grand scheme of things, WHAT IMAGES YOU TAKE is more important
then the order you take them
- We do on average 150-200 studies a day with
5 surgeons reading daily, and 5 radiologists reading a few times a month.
If we didn’t have our reporting system, we would 100% have a standard
image order or we would never function.
3- Timeframe
- This again depends on your u/s equipment,
ABI machine and software. Years ago the standard was 1.5hours for
an LEA but the older machines required pressing 5 buttons to change 1 thing
and 15 different tweaks to make a pretty image.
- We have all new
GE machines, newer parks machines (at majority of our locations), and techs
do a short H&P with a full detailed report at completion. We
perform ABI's and B/L Arterial with the completed report in 1 hour but
in some of our locations we have 1.5hrs for B/L due to dual techs and older
machines. Unilateral studies are always 1hour or less regardless
of the location/machine.
- I have worked
in labs with older machines and you do need that extra 30mins at times
for bilateral studies or difficult/challenging patients.
- For a Unilateral study (especially simplistic bypass graft
such as a fem-pop), they should not need more then 1 hour.
- I would recommend monitoring their studies. See if
they are using that entire 1.5 hours or if they are finishing early and
lounging for 30mins.
- All this aside, I have also, in the past,
worked in a VERY respectable lab with old HDI-5000 machines and we did
ABI's, B/L Arterial and AortoIliac all in 1 hour. Plus we did the
reporting. So I guess even by old practice standards I am reluctant
to really say 1.5hrs is 'needed'.
I hope this helps.
Andrew M. Bachman, Bs RVT
St. Lukes University Hospital and Health Network The Heart and Vascular
Center
"To handle yourself, use your head. To handle others, use your heart"
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]<mailto:[log in to unmask]>]
On Behalf Of Josie Klapec
Sent: Friday, December 23, 2016 10:15 AM
To: [log in to unmask]<mailto:[log in to unmask]>
Subject: DEVELOPED STANDARDS FOR IMAGING
Hello Everyone -
I posted this on the SVU site and got one response. I understand
that the questions are a bit basic but everyone's opinions are welcome.
I am updating protocols at a new job. They haven't been updated since
2001.
The technologists were taught to always measure vessels TRV x AP.
New students and myself (not new by any stretch of the imagination!)
have been taught to measure vessels (not masses) as AP x TRV. I
need to standardize and am wondering if this is an industry standard?
Also, the team has not had an effective supervisor for some time (4-5
years) Anarchy is rampant. Some scan the left first and others
scan the right. Once again, new students and myself have been taught
to always start on the right, (disregard extenuating circumstances). Any
standardization is met with a lot of criticism. I have repeated reproducibility,
surgeons wishes, etc.
Finally, I have gotten a lot of push back from them stating they are unable
to complete a unilateral fem-pop bypass graft and ABIs in one hour. How
much time do you allow?
I have reference both IAC and SVU standards/guidelines. This is not
a small lab - I have 22 direct reports. Any thoughts are greatly
appreciated!! Quite a bit to respond to, but please, if it is slow,
answer away!
Josie Klapec, BS, RVT
To unsubscribe or search other topics on UVM Flownet link to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__list.uvm.edu_archives_uvmflownet.html&d=DQIBaQ&c=kLuyfJs8xHieIsti43i5Bw&r=nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=EX5MA9x3D7W_q8SmeNCaBmdGYHFx3OYgY4gB2s5y62E&s=5XGuKWYeJIQGlMptsioOsUa71N5E8TiRWxZcpvTCMyI&e=
[St. Luke’s University Hospital is a 4-time recipient of this prestigious
award – Click the image to learn more]<http://www.slhn.org/top100>
Confidentiality Notice: This e-mail message, including any attachments,
is for the sole use of intended recipient(s) and may contain confidential
and privileged information. Any unauthorized review, use, disclosure or
distribution is prohibited. If you are not the intended recipient, please
contact the sender by reply e-mail and destroy all copies of the original
message.
To unsubscribe or search other topics on UVM Flownet link to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__list.uvm.edu_archives_uvmflownet.html&d=DQIDaQ&c=kLuyfJs8xHieIsti43i5Bw&r=nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=vD_EIjFKGl9cMgs4mwlhhWwzj8ldS5jKwRLNhLwEIY8&s=T-IUflS8MOd58Ea_lOwz8qZc0y0-UlgN7LWTDUM2QbY&e=
To unsubscribe or search other topics on UVM Flownet link to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__list.uvm.edu_archives_uvmflownet.html&d=DQIDaQ&c=kLuyfJs8xHieIsti43i5Bw&r=nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=vD_EIjFKGl9cMgs4mwlhhWwzj8ldS5jKwRLNhLwEIY8&s=T-IUflS8MOd58Ea_lOwz8qZc0y0-UlgN7LWTDUM2QbY&e=
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
To unsubscribe or search other topics on UVM Flownet link to: http://list.uvm.edu/archives/uvmflownet.html
To unsubscribe or search other topics on UVM Flownet link to:
------------------------------
Date: Tue, 27 Dec 2016 15:53:03 -0800
From: Steve Knight <[log in to unmask]>
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
You can try. It's sort for declaring in advance that you agree to disagree.
On Dec 27, 2016 3:51 PM, "Derek Butler" <[log in to unmask]>
wrote:
Steve –
Had to comment – I love this statement: “I know my stance on this is
not
the popular one but I believe my reasons are valid.” – I wonder if that
will work at home, lolol.
Derek -
*From:* UVM Flownet [mailto:[log in to unmask]]
*On Behalf Of *Steve
Knight
*Sent:* Tuesday, December 27, 2016 4:49 PM
*To:* [log in to unmask]
*Subject:* Re: DEVELOPED STANDARDS FOR IMAGING
Josie, when you say AP X TRV, do you mean AP in a longitudinal plane or
AP
in a transverse plane? AP in a transverse plane is prone to error due to
misalignment (particularly in the proximal aorta). TRV in the transverse
plane is also prone to error due to poor resolution of the lateral walls
and misalignment.
I like to use AP in a long plane. I will also measure in TRV for comparison
and they should be close but I trust the AP long more.
I know my stance on this is not the popular one but I believe my reasons
are valid.
~S
On Tue, Dec 27, 2016 at 10:01 AM, Bachman, Andrew <[log in to unmask]>
wrote:
Colleen,
Please email me on my work email and we can discuss this. We love
our
software but don’t want to start a software battle in this email chain.
[log in to unmask]
Andrew M. Bachman, Bs RVT
St. Lukes University Hospital and Health Network
The Heart and Vascular Center
"To handle yourself, use your head. To handle others, use your heart"
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Douville,
Colleen
Sent: Tuesday, December 27, 2016 12:52 PM
To: [log in to unmask]
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Andrew<
Which reporting software are you using?
Colleen Douville
Director l Vascular Ultrasound & Neurophysiology
[log in to unmask]
Office: 206-320-4080
Cell: 206-601-8077
Swedish Neuroscience Institute
500 17th Ave l 4 West, Room 400
Seattle l WA l 98122
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Bachman,
Andrew
Sent: Friday, December 23, 2016 8:02 AM
To: [log in to unmask]
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
Hi Josie,
I can not answer what the 'standard' is across the board because there
are
some labs that go above and beyond, some labs that do the bare minimum
and
hundreds in between. I work for the largest heart and vascular center
in
the country currently (by sheer number of locations and studies, and techs)
and I will answer based on our practices but we by no means set the
standards for our field.
1- Measuring:
- I have always seen vascular measuring AP
x TRV but this could be
a regional issue or a difference in whom is reading studies (Surgeons vs
Radiologists). We have had issues with dual general/vascular technologists
giving push back on this in years past.
- If everything is consistent and uniform across
the board with all
techs, this should not be an issue but 1 person differs and you will create
a ton of confusion to both ordering doctors and reading physicians.
2- Imaging order/standard
- This depends slightly on your reporting software.
- We use a software that allows our doctors
to click on a segment
of the artery and pull up only the images for that location allowing them
to see diseased areas side by side with previous images. In our case,
order is never a problem and techs are free to scan right then left, or
left then right, or any variation.
- If your physicians review each individual
image, then you will
need a standard across the board for everyone but I think in the grand
scheme of things, WHAT IMAGES YOU TAKE is more important then the order
you
take them
- We do on average 150-200 studies a day with
5 surgeons reading
daily, and 5 radiologists reading a few times a month. If we didn’t
have
our reporting system, we would 100% have a standard image order or we would
never function.
3- Timeframe
- This again depends on your u/s equipment,
ABI machine and
software. Years ago the standard was 1.5hours for an LEA but the
older
machines required pressing 5 buttons to change 1 thing and 15 different
tweaks to make a pretty image.
- We have all new
GE machines, newer parks machines (at
majority of our locations), and techs do a short H&P with a full detailed
report at completion. We perform ABI's and B/L Arterial with the
completed
report in 1 hour but in some of our locations we have 1.5hrs for B/L due
to
dual techs and older machines. Unilateral studies are always 1hour
or less
regardless of the location/machine.
- I have worked
in labs with older machines and you do need
that extra 30mins at times for bilateral studies or difficult/challenging
patients.
- For a Unilateral study (especially simplistic
bypass graft such as a fem-pop), they should not need more then 1 hour.
- I would recommend monitoring their studies. See
if they are using that entire 1.5 hours or if they are finishing early
and
lounging for 30mins.
- All this aside, I have also, in the past,
worked in a VERY
respectable lab with old HDI-5000 machines and we did ABI's, B/L Arterial
and AortoIliac all in 1 hour. Plus we did the reporting. So
I guess even
by old practice standards I am reluctant to really say 1.5hrs is 'needed'.
I hope this helps.
Andrew M. Bachman, Bs RVT
St. Lukes University Hospital and Health Network The Heart and Vascular
Center
"To handle yourself, use your head. To handle others, use your heart"
-----Original Message-----
From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Josie Klapec
Sent: Friday, December 23, 2016 10:15 AM
To: [log in to unmask]
Subject: DEVELOPED STANDARDS FOR IMAGING
Hello Everyone -
I posted this on the SVU site and got one response. I understand
that the
questions are a bit basic but everyone's opinions are welcome.
I am updating protocols at a new job. They haven't been updated since
2001.
The technologists were taught to always measure vessels TRV x AP.
New students and myself (not new by any stretch of the imagination!)
have been taught to measure vessels (not masses) as AP x TRV. I
need to standardize and am wondering if this is an industry standard?
Also, the team has not had an effective supervisor for some time (4-5
years) Anarchy is rampant. Some scan the left first and others
scan the
right. Once again, new students and myself have been taught to always
start on the right, (disregard extenuating circumstances). Any
standardization is met with a lot of criticism. I have repeated
reproducibility, surgeons wishes, etc.
Finally, I have gotten a lot of push back from them stating they are unable
to complete a unilateral fem-pop bypass graft and ABIs in one hour. How
much time do you allow?
I have reference both IAC and SVU standards/guidelines. This is not
a
small lab - I have 22 direct reports. Any thoughts are greatly
appreciated!! Quite a bit to respond to, but please, if it is slow,
answer
away!
Josie Klapec, BS, RVT
To unsubscribe or search other topics on UVM Flownet link to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__list.
uvm.edu_archives_uvmflownet.html&d=DQIBaQ&c=kLuyfJs8xHieIsti43i5Bw&r=
nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=EX5MA9x3D7W_
q8SmeNCaBmdGYHFx3OYgY4gB2s5y62E&s=5XGuKWYeJIQGlMptsioOsUa71N5E8T
iRWxZcpvTCMyI&e=
[St. Luke’s University Hospital is a 4-time recipient of this prestigious
award – Click the image to learn more]<http://www.slhn.org/top100>
Confidentiality Notice: This e-mail message, including any attachments,
is
for the sole use of intended recipient(s) and may contain confidential
and
privileged information. Any unauthorized review, use, disclosure or
distribution is prohibited. If you are not the intended recipient, please
contact the sender by reply e-mail and destroy all copies of the original
message.
To unsubscribe or search other topics on UVM Flownet link to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__list.
uvm.edu_archives_uvmflownet.html&d=DQIDaQ&c=kLuyfJs8xHieIsti43i5Bw&r=
nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=vD_
EIjFKGl9cMgs4mwlhhWwzj8ldS5jKwRLNhLwEIY8&s=T-IUflS8MOd58Ea_
lOwz8qZc0y0-UlgN7LWTDUM2QbY&e=
To unsubscribe or search other topics on UVM Flownet link to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__list.
uvm.edu_archives_uvmflownet.html&d=DQIDaQ&c=kLuyfJs8xHieIsti43i5Bw&r=
nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=vD_
EIjFKGl9cMgs4mwlhhWwzj8ldS5jKwRLNhLwEIY8&s=T-IUflS8MOd58Ea_
lOwz8qZc0y0-UlgN7LWTDUM2QbY&e=
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
------------------------------
Date: Tue, 27 Dec 2016 15:53:38 -0800
From: Steve Knight <[log in to unmask]>
Subject: Re: DEVELOPED STANDARDS FOR IMAGING
It sounds better than "I'm right and the rest of the world is wrong".
On Dec 27, 2016 3:51 PM, "Derek Butler" <[log in to unmask]>
wrote:
> Steve –
>
>
>
> Had to comment – I love this statement: “I know my stance on this
is not
> the popular one but I believe my reasons are valid.” – I wonder
if that
> will work at home, lolol.
>
>
> Derek -
>
>
>
>
>
> *From:* UVM Flownet [mailto:[log in to unmask]]
*On Behalf Of *Steve
> Knight
> *Sent:* Tuesday, December 27, 2016 4:49 PM
> *To:* [log in to unmask]
> *Subject:* Re: DEVELOPED STANDARDS FOR IMAGING
>
>
>
> Josie, when you say AP X TRV, do you mean AP in a longitudinal plane
or AP
> in a transverse plane? AP in a transverse plane is prone to error
due to
> misalignment (particularly in the proximal aorta). TRV in the transverse
> plane is also prone to error due to poor resolution of the lateral
walls
> and misalignment.
>
> I like to use AP in a long plane. I will also measure in TRV for
> comparison and they should be close but I trust the AP long more.
>
> I know my stance on this is not the popular one but I believe my reasons
> are valid.
>
>
>
> ~S
>
>
>
> On Tue, Dec 27, 2016 at 10:01 AM, Bachman, Andrew <
> [log in to unmask]> wrote:
>
> Colleen,
> Please email me on my work email and we can discuss this. We
love our
> software but don’t want to start a software battle in this email
chain.
>
> [log in to unmask]
>
> Andrew M. Bachman, Bs RVT
> St. Lukes University Hospital and Health Network
> The Heart and Vascular Center
>
> "To handle yourself, use your head. To handle others, use your
heart"
>
>
>
> -----Original Message-----
>
> From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Douville,
> Colleen
> Sent: Tuesday, December 27, 2016 12:52 PM
> To: [log in to unmask]
> Subject: Re: DEVELOPED STANDARDS FOR IMAGING
>
> Andrew<
> Which reporting software are you using?
>
>
> Colleen Douville
> Director l Vascular Ultrasound & Neurophysiology
> [log in to unmask]
> Office: 206-320-4080
> Cell: 206-601-8077
>
> Swedish Neuroscience Institute
> 500 17th Ave l 4 West, Room 400
> Seattle l WA l 98122
>
>
>
>
> -----Original Message-----
> From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Bachman,
> Andrew
> Sent: Friday, December 23, 2016 8:02 AM
> To: [log in to unmask]
> Subject: Re: DEVELOPED STANDARDS FOR IMAGING
>
> Hi Josie,
> I can not answer what the 'standard' is across the board because there
are
> some labs that go above and beyond, some labs that do the bare
minimum and
> hundreds in between. I work for the largest heart and vascular
center in
> the country currently (by sheer number of locations and studies, and
techs)
> and I will answer based on our practices but we by no means set the
> standards for our field.
>
> 1- Measuring:
> - I have always seen vascular measuring
AP x TRV but this could be
> a regional issue or a difference in whom is reading studies (Surgeons
vs
> Radiologists). We have had issues with dual general/vascular
technologists
> giving push back on this in years past.
> - If everything is consistent and uniform
across the board with
> all techs, this should not be an issue but 1 person differs and you
will
> create a ton of confusion to both ordering doctors and reading physicians.
> 2- Imaging order/standard
> - This depends slightly on your reporting
software.
> - We use a software that allows our doctors
to click on a segment
> of the artery and pull up only the images for that location allowing
them
> to see diseased areas side by side with previous images. In
our case,
> order is never a problem and techs are free to scan right then left,
or
> left then right, or any variation.
> - If your physicians review each individual
image, then you will
> need a standard across the board for everyone but I think in the grand
> scheme of things, WHAT IMAGES YOU TAKE is more important then the
order you
> take them
> - We do on average 150-200 studies a day
with 5 surgeons reading
> daily, and 5 radiologists reading a few times a month. If we
didn’t have
> our reporting system, we would 100% have a standard image order or
we would
> never function.
> 3- Timeframe
> - This again depends on your u/s equipment,
ABI machine and
> software. Years ago the standard was 1.5hours for an LEA but
the older
> machines required pressing 5 buttons to change 1 thing and 15 different
> tweaks to make a pretty image.
> - We have
all new GE machines, newer parks machines (at
> majority of our locations), and techs do a short H&P with a full
detailed
> report at completion. We perform ABI's and B/L Arterial with
the completed
> report in 1 hour but in some of our locations we have 1.5hrs for B/L
due to
> dual techs and older machines. Unilateral studies are always
1hour or less
> regardless of the location/machine.
> - I have worked
in labs with older machines and you do
> need that extra 30mins at times for bilateral studies or
> difficult/challenging patients.
>
- For a Unilateral study (especially simplistic
> bypass graft such as a fem-pop), they should not need more then 1
hour.
>
- I would recommend monitoring their studies. See
> if they are using that entire 1.5 hours or if they are finishing early
and
> lounging for 30mins.
> - All this aside, I have also, in the
past, worked in a VERY
> respectable lab with old HDI-5000 machines and we did ABI's, B/L Arterial
> and AortoIliac all in 1 hour. Plus we did the reporting. So
I guess even
> by old practice standards I am reluctant to really say 1.5hrs is 'needed'.
>
> I hope this helps.
> Andrew M. Bachman, Bs RVT
> St. Lukes University Hospital and Health Network The Heart and Vascular
> Center
>
> "To handle yourself, use your head. To handle others, use your
heart"
>
>
>
>
> -----Original Message-----
> From: UVM Flownet [mailto:[log in to unmask]]
On Behalf Of Josie
> Klapec
> Sent: Friday, December 23, 2016 10:15 AM
> To: [log in to unmask]
> Subject: DEVELOPED STANDARDS FOR IMAGING
>
> Hello Everyone -
>
> I posted this on the SVU site and got one response. I understand
that the
> questions are a bit basic but everyone's opinions are welcome.
>
> I am updating protocols at a new job. They haven't been updated
since
> 2001.
>
> The technologists were taught to always measure vessels TRV x AP.
> New students and myself (not new by any stretch of the imagination!)
> have been taught to measure vessels (not masses) as AP x TRV.
I
> need to standardize and am wondering if this is an industry standard?
>
> Also, the team has not had an effective supervisor for some time (4-5
> years) Anarchy is rampant. Some scan the left first and
others scan the
> right. Once again, new students and myself have been taught
to always
> start on the right, (disregard extenuating circumstances). Any
> standardization is met with a lot of criticism. I have repeated
> reproducibility, surgeons wishes, etc.
>
> Finally, I have gotten a lot of push back from them stating they are
> unable to complete a unilateral fem-pop bypass graft and ABIs in one
hour.
> How much time do you allow?
>
> I have reference both IAC and SVU standards/guidelines. This
is not a
> small lab - I have 22 direct reports. Any thoughts are greatly
> appreciated!! Quite a bit to respond to, but please, if it is
slow, answer
> away!
>
> Josie Klapec, BS, RVT
>
> To unsubscribe or search other topics on UVM Flownet link to:
> https://urldefense.proofpoint.com/v2/url?u=http-3A__list.
> uvm.edu_archives_uvmflownet.html&d=DQIBaQ&c=kLuyfJs8xHieIsti43i5Bw&r=
> nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=EX5MA9x3D7W_
> q8SmeNCaBmdGYHFx3OYgY4gB2s5y62E&s=5XGuKWYeJIQGlMptsioOsUa71N5E8T
> iRWxZcpvTCMyI&e=
>
> [St. Luke’s University Hospital is a 4-time recipient of this prestigious
> award – Click the image to learn more]<http://www.slhn.org/top100>
>
> Confidentiality Notice: This e-mail message, including any attachments,
is
> for the sole use of intended recipient(s) and may contain confidential
and
> privileged information. Any unauthorized review, use, disclosure or
> distribution is prohibited. If you are not the intended recipient,
please
> contact the sender by reply e-mail and destroy all copies of the original
> message.
>
> To unsubscribe or search other topics on UVM Flownet link to:
>
> https://urldefense.proofpoint.com/v2/url?u=http-3A__list.
> uvm.edu_archives_uvmflownet.html&d=DQIDaQ&c=kLuyfJs8xHieIsti43i5Bw&r=
> nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=vD_
> EIjFKGl9cMgs4mwlhhWwzj8ldS5jKwRLNhLwEIY8&s=T-IUflS8MOd58Ea_
> lOwz8qZc0y0-UlgN7LWTDUM2QbY&e=
>
> To unsubscribe or search other topics on UVM Flownet link to:
> https://urldefense.proofpoint.com/v2/url?u=http-3A__list.
> uvm.edu_archives_uvmflownet.html&d=DQIDaQ&c=kLuyfJs8xHieIsti43i5Bw&r=
> nowbzADbgbSQGUo4tYN6uL7V4skFKveH2RQh4AiNuKs&m=vD_
> EIjFKGl9cMgs4mwlhhWwzj8ldS5jKwRLNhLwEIY8&s=T-IUflS8MOd58Ea_
> lOwz8qZc0y0-UlgN7LWTDUM2QbY&e=
>
>
> To unsubscribe or search other topics on UVM Flownet link to:
> http://list.uvm.edu/archives/uvmflownet.html
>
>
>
> To unsubscribe or search other topics on UVM Flownet link to:
> http://list.uvm.edu/archives/uvmflownet.html
> To unsubscribe or search other topics on UVM Flownet link to:
> http://list.uvm.edu/archives/uvmflownet.html
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
------------------------------
End of UVMFLOWNET Digest - 24 Dec 2016 to 27 Dec 2016 (#2016-190)
*****************************************************************
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html
This email is subject to certain disclaimers, which may be reviewed via
the following link: http://apps01.prominic.net/A55CF6/MSOIndex.nsf/DisclaimerX.xsp
To unsubscribe or search other topics on UVM Flownet link to:
http://list.uvm.edu/archives/uvmflownet.html