Hi Eduardo,

Its the grounding wire for the 24V power distribution board in the back of the instrument.  The wire in older ECS is too small and results in exactly this issue (instant temp drop, overshoot temp, machine goes “not ready").  I will send Costech's instructions and pictures off list.  Fixed ours and other ECS.  Easy to do.  Jason

Jason Curtis, Ph.D.
Stable Isotope Mass Spec Lab manager
Senior Associate-In Geochemistry
Department of Geological Sciences
University of Florida
Gainesville, FL 32611
+1-352-275-8642




On Sep 4, 2018, at 3:36 PM, Dias de Oliveira, Eduardo A. <[log in to unmask]> wrote:

Hi Mike, no problem with that too, we got exactly 208v
Thanks!



On Sep 4, 2018, at 2:16 PM, Mike Elefante, Elf Analytical Services <[log in to unmask]> wrote:

This happened after a power outage, check your incoming power. If it is below 208vac I would suggest boosting it with a transformer.


Mike Elefante, President
Elf Analytical Services, Inc
732-922-8400
 



-----Original Message-----
From: Dias de Oliveira, Eduardo A. <[log in to unmask]>
To: ISOGEOCHEM <[log in to unmask]>
Sent: Tue, Sep 4, 2018 3:13 pm
Subject: Re: [ISOGEOCHEM] EA

Hi Andrew,
Yes, new thermocouple.
Yes, talked to Bruno and John, and so far we narrowed down to that to rule out grounding issues.
Thanks.


On Sep 4, 2018, at 2:09 PM, Andrew Schauer <[log in to unmask]> wrote:

Hi Eduardo,

Did you replace the thermocouple as well? Does the second column also vary in temperature? Have you asked Bruno or Jon at Costech?

andy

On Tue, Sep 4, 2018 at 12:05 PM Dias de Oliveira, Eduardo A. <[log in to unmask]> wrote:
Hi everyone, 
Our EA (Costech) is funny since some power outage we had a few months ago.
When the machine clicks to turn the carousel for a new sample, the temperature drops (in less than one second) from 1020 C to around 1000 CO2 Then it overshoots to around 1030 to compensate and oscillates until stable at 1020 again. I got new candles, all grounds in place and now I think my last and worse place to investigate is the board. Does anyone have seen anything similar?
Does anyone have any tips? Is it worth to change the board? 
More: Is it worth to replace the 17 year old EA? Should I get a thermo or what do you recommend? I run plant, animals and soils for C and N mostly. 
Thank very much, 
I appreciate any input.
Cheers

Eduardo Dias de Oliveira
+1(312)996 5929
Postdoctoral Research Associate
Department of Biological Sciences
University of Illinois at Chicago



--

Eduardo Dias de Oliveira
Postdoctoral Research Associate
Department of Biological Sciences
University of Illinois at Chicago




Eduardo Dias de Oliveira
Postdoctoral Research Associate
Department of Biological Sciences
University of Illinois at Chicago